Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

error code F0D0A for SONOS 1000

Oral Unal

error code F0D0A for SONOS 1000

August 17, 2004 11:25

Error code F0D0A seen on seven segment display SONOS 1000 Echo and some time CF973.

Could anybody help me?

Sincerely

Log inor Register

to rate and post a comment

Jim Carr

Re: error code F0D0A for SONOS 1000

August 18, 2004 10:12

Rick Reed, our systems engineer, advises:
F0D0A points to a total communication failure between the "controller"(located with the monitor) and the Scan Convertor. Could be lack of power to Scan Convertor (check round connectors on back of e-box) or comm path failure over the HPIB cable bus.
Jim Carr, Sonora Medical Systems

Log inor Register

to rate and post a comment