Error 16 GE 600T
March 25, 2006 08:50
I have error 16 in Ge 600T when the unit is on, and I can not selec focus 0.3 only smal is work.
Could some body tell me what is the problem??
to rate and post a comment
Manuel Casanova
Re: Error 16 GE 600T
March 25, 2006 09:08
cause 1: The heater current board 400pl1, or heater power supply 400pl2 are not in position.
action 1: insert the board in slots.
cause 2:the large spot heater is not working properly
action 2: run the heater test job card lg004
cause 3 and just cause: the tube filament large is out,
Action 3:Replace the XRay tube.
bye
to rate and post a comment
Joel Inacio
Re: Error 16 GE 600T
September 02, 2006 07:53
Insert a lamp (12v) between ground and small of hv cable. If the code error desapear the problem is in the tube.
to rate and post a comment
Frank Beltre
Re: Error 16 GE 600T
February 05, 2007 10:02
Yojin,
I lost your email. Can you send it to me again.
Frank Beltre