Elscint Select CT

August 26, 2010 10:29

The ignition sequence takes too many times and sometimes never ends. What is the problem? CPI? or RMS?

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

Re: Elscint Select CT

September 08, 2010 03:47

Long "ignition" time can be caused by problems with "Beam forming".
Open SCU-diag window (click on proper icon which is under application window). Then you can see a messages from SCU.
Check if you have messages begins with PT80,....
For example PT80,FFFD can be caused by timeout with proper thickness setup, PT80,FFFE - done by wrong compensator state during air calibration).
Beam unit is controlled by SCU and sometimes it is a problem with SCU, sometimes with Beam mechanics.
Check which messages you get and send me an info for diagnostics
regards
Wojciech