We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.
Abu-Yusuf
DRYPIX 2000 Dicom Printer
October 20, 2016 10:47
Hi everybody, i have problem with drypix 2000. I did not find Set Network menu in it. Menu which i got is followings:
0.RETURN
1.SET DATE
2.CHECK VERSION
3.(does not exist in my printer)
4.(does not exist in my printer)
5.QC
6.SLEEP MODE
7.CALIBRATION
i aslo tried to connect with its default AE TITLE and IP address (172.16.1.30 , DRYOPIX , 17238) but it does not help.
How can i do factory reset. Or any idea?
Thanks a lot.
to rate and post a comment
Dr_Danyashi
re: DRYPIX 2000 Dicom Printer
October 21, 2016 09:19
Try setting a different IP address and set same to your workstation. reboot system and check the menu again. Hope this will help
to rate and post a comment
Gabriel Cool
re: DRYPIX 2000 Dicom Printer
October 21, 2016 09:20
You need enter in M utility. For this you must push two buttoms first and last(4). 1bottom+4bottom push for 3 seconds and then you will have full acsess for settings
to rate and post a comment
Lenny Place
re: DRYPIX 2000 Dicom Printer
October 21, 2016 09:21
Are you in the user mode or maintenance mode? You have to enter the M-Mode to set these and if you are then it sounds life the compact Flash is corrupt and this will require using the PC Tool.
to rate and post a comment