Over 500 Total Lots Up For Auction at One Location - NJ 06/15

drager anesthesia fabius

(6)

Erick babadilla

drager anesthesia fabius

July 28, 2017 09:01

please help me to solve this problem.
our ventilator drager fabius anesthesia machine display on the monitor VENTILATOR INOP

Log inor Register

to rate and post a comment

(36)
(7)

Ray Brown

re: drager anesthesia fabius

July 30, 2017 08:43

Several things can be wrong. First, did it pass the Leak and Compliance test? Make sure all your connections are tight, check the roller diaphragm for cuts or holes, do a leak test of the vacumn system, then check both diaphrams again.

Log inor Register

to rate and post a comment

nimpus

re: drager anesthesia fabius

July 31, 2017 10:35

The last time I had this problem, it was causing the leaks. Performs the leak test and verifies the leakage of the system to the patient circuit. I hope you will solve the problem. Good to you

Log inor Register

to rate and post a comment

nimpus

re: drager anesthesia fabius

July 31, 2017 10:35

The last time I had this problem, it was causing the leaks. Performs the leak test and verifies the leakage of the system to the patient circuit. I hope you will solve the problem. Good to you

Log inor Register

to rate and post a comment

(4)

JOEL JOELPALZA

re: drager anesthesia fabius

August 25, 2017 12:11

Hi Erik, what fabius do you have? Maybe I have the manual.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Altima - Your CT / MRI Diagnostic Imaging Specialists!

Altima will inspire the trust, loyalty and commitment of our customers through positive relationships and collaboration, top quality equipment services, and a clearly superior value proposition. Call Today (469) 644-7063

DOTmed Certified stats

Próximos eventos de leilão

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)