CT/e detector Power Supply

May 16, 2012 09:31

Can somebody help me locate detector power supply in GE CT/e? Sometimes,I am getting several horizontal lines on scout image & a large ring artifact outside the image area. The display window also goes off scale. The problem disappears after switching off the complete supply to the PDU & switching On again. Sometimes we have to switch OFF & ON PDU supply several times to get rid of this problem. The machine is a 200V power supply.
I presume it is the detector power supply, which may be posing this problem.
Any help would be greatly appreciate.

2 people liked this
Log inor Register

to rate and post a comment

(10)

daniel dougert

re: CT/e detector Power Supply

May 16, 2012 11:11

what machine. problem could be the das power supply.

Log inor Register

to rate and post a comment

(11)
(2)

A Mishra

re: re: CT/e detector Power Supply

May 16, 2012 02:15

I think he is looking for DAS power supply in Ge CT/e.

Log inor Register

to rate and post a comment

(11)
(2)

A Mishra

re: CT/e detector Power Supply

May 16, 2012 02:15

I think you are looking for DAS power supply.

Log inor Register

to rate and post a comment

(31)
(1)
(12)

S Goyal

re: CT/e detector Power Supply

May 17, 2012 12:46

It is located just behind the detector on the slip ring. It is in a perforated metal box.

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: CT/e detector Power Supply

May 17, 2012 12:47

There are other possibalities that caused this problem except Power Supply.
1, please Check the Fans
2, DAS board or detector.

Log inor Register

to rate and post a comment

(11)
(2)

A Mishra

re: re: CT/e detector Power Supply

May 17, 2012 09:31

Which fans, please adv. so that we can also understand about this problem?
What to look for in DAS boards or detector?

Log inor Register

to rate and post a comment

(11)
(2)

A Mishra

re: re: CT/e detector Power Supply

May 17, 2012 09:32

Dear Mr. Lin,
Plz adv how the problem disappears after switching Off PDU & then switching ON.

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: re: re: CT/e detector Power Supply

May 17, 2012 01:08

yes,
the problem maybe come like this way:
When the Fan for the detector board( CAM BOARD)is bad or dirty, the CAM borads will get hot, this will make some artifact for the image.When you switch off the power, it will cool down, and back to normal.
What is your phone? I call you for this.

Log inor Register

to rate and post a comment

S.DHINESH RAJ

re: re: re: re: CT/e detector Power Supply

May 18, 2012 09:27

mr.lin..
this prblm som time coming nprm board also....... and flat cables dirty also

Log inor Register

to rate and post a comment

S.DHINESH RAJ

re: re: re: re: CT/e detector Power Supply

May 18, 2012 09:27

mr.lin detector may be possible

Log inor Register

to rate and post a comment

S.DHINESH RAJ

re: re: re: re: CT/e detector Power Supply

May 18, 2012 09:25

mr lin.... iam sitting one prblm of ct/e my system disk going to crash after done 5,6 patient... already i change o2 sytem module,dpbci board,nprm,nprif,back plane,scsi, my sytem connect wit ups only so no power r ground prblm wats ur opinion......

Log inor Register

to rate and post a comment

(31)
(1)
(12)

S Goyal

re: re: re: re: re: CT/e detector Power Supply

May 18, 2012 10:33

Check memory.

Log inor Register

to rate and post a comment

S.DHINESH RAJ

re: re: re: re: re: re: CT/e detector Power Supply

May 18, 2012 11:06

mr.goyal ji this ans for whom..... and how to chek memory?

Log inor Register

to rate and post a comment

S.DHINESH RAJ

re: re: re: re: re: re: CT/e detector Power Supply

May 18, 2012 11:06

goyal ji.... i install new sys disk and raw data disk memory is emty......

Log inor Register

to rate and post a comment

(31)
(1)
(12)

S Goyal

re: re: re: re: re: re: re: CT/e detector Power Supply

May 18, 2012 12:05

Try by changing DIM. We have seen problem due to DIM.

Log inor Register

to rate and post a comment

S.DHINESH RAJ

re: CT/e detector Power Supply

May 18, 2012 10:33

dear,
confirm somethings......
1)any err is there?
2)wats the radius of ring artifact?
3)your cosoule is connected with ups r pdu?
plz send image to i tell u the soulution sdrajct@gmail.com
this type prblm due to sometime nprm board also.......and flat cables,camboards,cif,ddp...... plz send the image iwill confirm u.....

Log inor Register

to rate and post a comment

stats
DOTmed text ad

OEM Philips & Siemens parts with Warranty!

Email P/N for pricing: sales@Relaxationinc.com (P) 813-925-1314, www.relaxationinc.com

stats

Próximos eventos de leilão

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)