We have just changed the x-ray tube and the tank was changed last December. We still have this message "The tube anode voltage (anode output from the HV tank) measured was over specifications".
May I have your technical idea on the problem that this system GE Sytec SRI could have?

Thanks a lot

Log inor Register

to rate and post a comment

(86)
(5)

SHIJU KARAKKOOTTIL

Re: CT GE Sytec SRI Error

September 25, 2007 03:43

Hi,
Check the Anode HV cable. You can interchange the Anode, cathode HV cables and check.
Perform Anode/cathode High voltage test (with tube and without tube) Be carefull while doing the test. Instructions are availble at Diagnosics Manual.
Perform KV Board test.
Check the machine Ground voltage.
Not down the error LED s displaying at KV Board.
Mail me still you have problem.
All the best,

Shiju
Director
Cura Medical Equipment Pvt Ltd
India