Error is:
E-IOC_ERACTL_OPENRELAIS
Reason: IOBUS control module: Erasure control board safety test:
Safety Relais does not open again.
Meaning: Safety relais on Erasure control board for interrupt power supply
to erasure lamps does not open.
Hardware Problem.
Cure: Replace Erasure control Board.
Send me an e-mail and I'll send you the service manual for the 85 and IMOS Show Error Viewer.
rswapp@comcast.net
Authorized AGFA Field Service Tech.
My guess is you lost the Erasure Control Board, GS102 number: CM+9515564204.
Or you may have a problem with the Current Sense Board, GS104. part number: CM+9515564401.
Go for the GS102 First, I believe there is a micro relay that has a contact that is welded closed from over current.
One call to 440-263-6575 and your move will be covered.
We repair & store trailers. We build to your custom needs.
JDI Mobile Medical Manufacturing. Inc. We have your back.
Korvin Kel
CR 85-X
February 13, 2018 08:08
Hello.
Can someone help me with an explanation of the error SERVICE 24724? Tnx.
to rate and post a comment
Ronald Swapp
re: CR 85-X
February 13, 2018 04:40
Error is:
E-IOC_ERACTL_OPENRELAIS
Reason: IOBUS control module: Erasure control board safety test:
Safety Relais does not open again.
Meaning: Safety relais on Erasure control board for interrupt power supply
to erasure lamps does not open.
Hardware Problem.
Cure: Replace Erasure control Board.
Send me an e-mail and I'll send you the service manual for the 85 and IMOS Show Error Viewer.
rswapp@comcast.net
Authorized AGFA Field Service Tech.
to rate and post a comment
Ronald Swapp
re: CR 85-X
February 13, 2018 04:40
My guess is you lost the Erasure Control Board, GS102 number: CM+9515564204.
Or you may have a problem with the Current Sense Board, GS104. part number: CM+9515564401.
Go for the GS102 First, I believe there is a micro relay that has a contact that is welded closed from over current.