Greeting Thomas the code 384 is:
Description: POST HVDC to30V failed- supply too low
Possible Cause: The HVPS output is too low during POST.
Check: Cycle system power and determine if this error is repeated. Watch LED2 on the HVPS during
POST and confirm that it illuminates. If not, disconnect the cable between the HVPS and the RF Amplifier and cycle whileagain watching LED2. If it is now illuminated, a problem with the RF Amplifier is likely. Configure system for test mode operation. Verify HVPS control signals from the controller are correct while activated. Verify the HVPS is functioning correctly.
Correction: Verify that all cables are properly connected.
Replace the Controller assembly.
Replace or repair the HVPS.
Replace or repair theRF Amplifier.
We have service manual, please send your email and we send you it.
Alvaro
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Thomas Marr
Conmed 5000 error code 384
October 02, 2011 11:29
Hello,
Does anyone have information on error code
384 on a 5000 Conmed , Waiting on service manual
Doctor having a tantrum.
Thanks
Thomas
to rate and post a comment
Alvaro Sandoval
Re: Conmed 5000 error code 384
October 02, 2011 06:27
Greeting Thomas the code 384 is:
Description: POST HVDC to30V failed- supply too low
Possible Cause: The HVPS output is too low during POST.
Check: Cycle system power and determine if this error is repeated. Watch LED2 on the HVPS during
POST and confirm that it illuminates. If not, disconnect the cable between the HVPS and the RF Amplifier and cycle whileagain watching LED2. If it is now illuminated, a problem with the RF Amplifier is likely. Configure system for test mode operation. Verify HVPS control signals from the controller are correct while activated. Verify the HVPS is functioning correctly.
Correction: Verify that all cables are properly connected.
Replace the Controller assembly.
Replace or repair the HVPS.
Replace or repair theRF Amplifier.
We have service manual, please send your email and we send you it.
Alvaro
to rate and post a comment
Thomas Marr
Re: Conmed 5000 error code 384
October 17, 2011 10:09
Dear Alviro
Thanks so much oure-mail is arrowheadbio@charter.net
Thomas