BV pulsera

February 07, 2020 10:25

system will not shutdown, stays in shutdown screen. When plugged in powers up on own, replaced on/off pcb and switch panel.

Log inor Register

to rate and post a comment

(82)
(15)

Herrmann

re: BV pulsera

February 10, 2020 09:18

Hi,
What for a Pulsera you have ? (718***)

And it gives a option DVD recorder. When this option
is in the software konfigured, then the shutdown-Screen
comes for about 45 sekunden, so that the DVD recoder
can finish the recordning.

But when you have the shutdown screen / welcom scree
when you switch on the system, then you have a software
problem and you must reload the software.

Kind regards

Log inor Register