ATL 5000 error # 002

Tags: Ultrasound, ATL

muhammad shahid

ATL 5000 error # 002

September 15, 2014 09:16

machine reboot and given this error someone help of this error. thanx and best reqards

Log inor Register

to rate and post a comment

(230)
(35)
(4)

Rey Gonzalez

re: ATL 5000 error # 002

September 15, 2014 06:19

removed the probes from the system, specially if you have a L12-5 and reboot the system, then plug the lineal probe in the center connector and try. for more help contact me rey@allultrasoundservices.com

Log inor Register

to rate and post a comment

(102)
(39)
(1)

Konstantin Kuznetsov

re: ATL 5000 error # 002

September 15, 2014 06:22

"0002 FEC alert message -
The front end has stopped scanning because of some power monitor problem. This could be a hardware problem with the scanhead, the channel boards, the AIM, or the power supplies, but this also could be a system software problem. In reviewing the previous error messages, from the FEC, the board errorID should indicate if there is a specific problem with hardware."

Manuals to your scanner are available on the Internet ;-). Look for it as PHILIPS HDI-5000. Google to you in the help.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.

stats
advertisement

Próximos eventos de leilão

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)