ATL 1000 ultrasound's power supply went out.
After replacing it, machine worked for 24 hours.
Then Error 4 appeared and it will not recognize the probes.
Machine was donated to Armenia, they do not have a service manual.
Check the voltages in the test points of the HV board (The expected voltages are labeled) you have to remove the probes port on the front and the board is in the left side
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
Bahram Hashemi
ATL 1000 Error 4
October 01, 2012 10:01
ATL 1000 ultrasound's power supply went out.
After replacing it, machine worked for 24 hours.
Then Error 4 appeared and it will not recognize the probes.
Machine was donated to Armenia, they do not have a service manual.
Any suggestion would be appreciated
to rate and post a comment
Ricardo A. Sanchez
re: ATL 1000 Error 4
October 01, 2012 05:51
send me an email and I will send you the service manual
to rate and post a comment
Bahram Hashemi
re: re: ATL 1000 Error 4
October 02, 2012 09:47
accuware1@yahoo.com
to rate and post a comment
wang jia
re: ATL 1000 Error 4
October 01, 2012 07:34
hi;
I have the service manuals of HDI1000,HDI4000,HDI5000,UM9HDI,HD3,HD7,HD11,iu22,ie33 etc.E-mail:sdcwcywjj@163.com.
to rate and post a comment
wang jia
re: ATL 1000 Error 4
October 01, 2012 07:34
send me an email and I will send you the service manual
to rate and post a comment
Bahram Hashemi
re: re: ATL 1000 Error 4
October 02, 2012 09:47
accuware1@yahoo.com
to rate and post a comment
Luis Vargas Carrillo
re: ATL 1000 Error 4
October 01, 2012 10:04
Check the voltages in the test points of the HV board (The expected voltages are labeled) you have to remove the probes port on the front and the board is in the left side
to rate and post a comment
Bahram Hashemi
re: re: ATL 1000 Error 4
October 02, 2012 09:47
Thank you so much
to rate and post a comment