Hello!
Help solve the problem with the crescent-shaped artifact on Aquilion32
On the scan it is presented in the form of a white strip.
Appears after some time of work.
If you turn off the computer and turn it on again, the artifact disappears
Please did you check DAS power supply- Detector Temperature- Air calibration must not be old. Please can you test both reconstruction rack ? Also check if a red led lit on SCRT or GCIF PWB when artifact appears ( this system is MUDAT or SureCom?
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
evg evgen
Aquilion32 Help!
April 14, 2017 05:50
Hello!
Help solve the problem with the crescent-shaped artifact on Aquilion32
On the scan it is presented in the form of a white strip.
Appears after some time of work.
If you turn off the computer and turn it on again, the artifact disappears
to rate and post a comment
QuantaMed
re: Aquilion32 Help!
April 17, 2017 04:43
Please did you check DAS power supply- Detector Temperature- Air calibration must not be old. Please can you test both reconstruction rack ? Also check if a red led lit on SCRT or GCIF PWB when artifact appears ( this system is MUDAT or SureCom?
to rate and post a comment