Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Aquilion 16 errors

Tags: CT, TOSHIBA

Posted by an Anonymous User

Aquilion 16 errors

September 02, 2012 08:02

In Aquilion 16 spellman generator, we have Errors HFG error & over ma(fixed), Can any body knows, what could be the problem?

Thanks for help

Log inor Register

to rate and post a comment

(53)
(11)
(16)

Dominique Bultot

re: Aquilion 16 errors

September 03, 2012 11:09

Have you connected your laptop with the software XC test ?

Log inor Register

to rate and post a comment

(25)
(6)
(11)

Nabil Naji

re: re: Aquilion 16 errors

September 03, 2012 07:40

Yes, I have connected my laptop and did exposure test and in the Error description gives me those errors.
During normal operation, the warm up can not be done and always gives error "over ma"

Log inor Register

to rate and post a comment

(39)
(5)
(1)

Jose Soto

re: re: re: Aquilion 16 errors

September 03, 2012 08:52

check if some arc noise is present send error code to me ...kathzinstrumentos@hotmail.com

Log inor Register

to rate and post a comment

(25)
(6)
(11)

Nabil Naji

re: re: re: re: Aquilion 16 errors

September 04, 2012 03:32

Hi Jose,

Please tell me how I can check if some arc noise is present?

The error code is : BFDFFFFFFF

Log inor Register

to rate and post a comment

Yuriy Kuznetsov

re: re: re: re: re: Aquilion 16 errors

September 05, 2012 09:58

Check the X-ray
exposure waveform
using the Signal Show
function.

Log inor Register

to rate and post a comment

(25)
(6)
(11)

Nabil Naji

re: re: re: re: re: re: Aquilion 16 errors

September 05, 2012 12:41

In the signal show,I can see that the ma signal is very high and more then 600ma and the kv signal is very low and around 1kv.
Any indication?

Log inor Register

to rate and post a comment

(26)
(3)

QuantaMed

re: re: re: re: re: re: re: Aquilion 16 errors

May 24, 2015 10:04

Hi

This error is due a high current inside of the xray tube. So , defective part is the xray tube. Please, replaced It.

Log inor Register

to rate and post a comment

(3)

Oleg Kuchin

re: Aquilion 16 errors

September 07, 2012 11:42

Its like Tube is broken. How long it working?

Log inor Register

to rate and post a comment

(25)
(6)
(11)

Nabil Naji

re: re: Aquilion 16 errors

September 09, 2012 09:53

The tube worked two years with only 170,000 slices.

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: re: re: Aquilion 16 errors

October 04, 2012 10:51

Dear Nabil, i`m having the same problem you have, could you tell me what was your conclusion? was the tube? i have the same slices number, 171 k

Log inor Register

to rate and post a comment

(88)
(36)

Anton Spaci

re: Aquilion 16 errors

October 05, 2012 10:04

Hi
It looks like tube problem
Doesn't mind how many slices it has
Depends how you use it , i got in my area same problem and some tube has also 500 000 slices or more
Concerning your problem , if I was to your place , i will check connection on generator and tube ,redo all connection with grease and oil/grease (depend on HV cable you use ) for the tube
If after this you get the same problem , i think you have to order some parts
best regards
anton

Log inor Register

to rate and post a comment

(42)
(14)

Pawel Kopysc

re: Aquilion 16 errors

May 06, 2015 03:08

Hi,

I have the same errors. Can You tell me what finally was broken?? Was it the tube?

Please email me kindly
kontakt@medikol.eu

Regards
Pawel

Log inor Register

to rate and post a comment

(29)
(8)

ALSELWI

re: re: Aquilion 16 errors

May 19, 2015 01:49

It is the Tube.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS Mx-8000 IDT/Brilliance 16 Slice CT Scanner -----Worldwide leader!!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Próximos eventos de leilão

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)