Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

amx4 45b error

(4)

(1)

dave wilson

amx4 45b error

July 28, 2011 12:40

replaced rotor bd and fil/kvp bd from previous problem shot xray left site now 45b error whats going on,thanks

Log inor Register

to rate and post a comment

Yuriy Kuznetsov

Re: amx4 45b error

July 28, 2011 07:49

ERROR 45B
FIL/KVP Board May Be Defective

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

Re: amx4 45b error

July 28, 2011 08:14

As yuri said, FIl/KV PCB might be defective. Also do a calibration of the Generator. Check HT cables or finally the big one the Xray tube might be acting up.