AGFA DRYSTAR 300 error MDM sensor

(62)
(11)
(14)

James Fower

AGFA DRYSTAR 300 error MDM sensor

March 20, 2017 05:32

Hi colleagues,
my customer is getting error:
LAST MDM-SENS NOT OK
Any help is appreciated. Service manual, discks
Thank you all in advance

1 person liked this
Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: AGFA DRYSTAR 300 error MDM sensor

March 22, 2017 10:24

MDM stands for"MACRODENSITOMETER". It is built-in to allow for constant image quality concerning Dmin, Dmax, sensitometry and vertical banding.

There are no adjustment procedures for the MDM but I would recommend you perform a full calibration of the MDM.

I can help with service manual and service software.
craig@telemed-technologies.com

Log inor Register

to rate and post a comment

(62)
(10)

Ricardo Rivera

re: AGFA DRYSTAR 300 error MDM sensor

March 24, 2017 10:36

hi, the MDM is the densitometer, you can calibrate it in the control panel to level 2 or 3 maybe it everything, if not check the halogen lamp of the mdm, maybe is bad or not clean, it is in the right side of the machine.