Agfa cr-35

July 04, 2016 10:34

Dear all :

I have two errors at agfa cr-35 (247AB, 20505) .
What these errors means?and how i can cancel it?

Pls any one can help me?

Best regards

Log inor Register

to rate and post a comment

(23)
(3)

Luis Espinoza

re: Agfa cr-35

July 04, 2016 06:54

Send me an email. Luismanuelespinoza@hotmail.es

Log inor Register

to rate and post a comment

(34)
(2)

sara ahmed

re: re: Agfa cr-35

July 05, 2016 03:01

Sara.ahmed.hazem@gmail.com

Log inor Register

to rate and post a comment

(2)

Mantas Mickus

re: Agfa cr-35

July 04, 2016 06:55

247AB
E_HDC_IP_JAM_TRA_VACUUM
Meaning : IP jam in transport unit.
Reason : Transport unit lost the image plate while transport of IP from
cassette down to scanner. No more vacuum is detected.
Cure : Open right side cover and put IP into cassette.

20505
E_SCN_LASER_ALARM
Meaning : Laser power after the warming-up phase is still below or equal
las_pow_alarm. Machine status is "locked".
Reason : - old laser with too low power
- bad (dusty) optics
- pd_factor not correct
Cure : Confirm message on user terminal and continue operation.
In the long term, replace the laser, if it is bad.

Log inor Register

to rate and post a comment

(34)
(2)

sara ahmed

re: re: Agfa cr-35

July 17, 2016 04:07

Dear mr.mantas
How can i open laser unit and clean it?
What (pd_factor not correct) is meaning?
Best regards

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: re: re: Agfa cr-35

July 20, 2016 10:32

you can check laser power history and status logging in web page of cr35. Enter by navegator with machine IP

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Source for Sony Medical Displays, Printers & More!

Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.

stats

Próximos eventos de leilão

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)