Some one knows the meaning of Error 22862 during self test on Agfa Compact Plus Digitizer? or can send me a copy of service manual.my email is doctoromair@yahoo.com ,Thanks you in advance for help.
re: re: re: AGFA Compact Plus Error 22862 on selftest
October 06, 2012 12:52
in my opinion
The problem is causinf from 3 factor as below:
1. Control PCB for Rotation is failed
2. Light Barrier sensor of Control PCB Rotation failed.
3. Light Barrier of A Cassette module is failed
try to clean all these above sensor and check again
Good Luck to you
Bao
Online auction of modern high-end laboratory and biopharma research equipment from the recent bankruptcy of Spexis AG in Allschwil near Basel, Switzerland. Start bidding today! closes on June 4. info@dome-auctions.com..+31(0)20 808 2880
Omair nawab
AGFA Compact Plus Error 22862 on selftest
February 03, 2012 12:22
Some one knows the meaning of Error 22862 during self test on Agfa Compact Plus Digitizer? or can send me a copy of service manual.my email is doctoromair@yahoo.com ,Thanks you in advance for help.
to rate and post a comment
Bao Vo Quoc
re: AGFA Compact Plus Error 22862 on selftest
June 04, 2012 09:25
Error Code --> 22862
E_ROT_AINIT_TOROTHOMESLOW,
Reason : Timeout while stepper drive movement 'rotation home positioning',
in rot_a_initialize() Rotation Module command function.
Meaning :
Cure :
ected.
- fuse blown.
- power supply defective.
Cure : - check cables and connectors.
- check/replace fuse.
- check/replace power supply.
Goodluck to you
Bao
to rate and post a comment
Omair nawab
re: re: AGFA Compact Plus Error 22862 on selftest
June 04, 2012 07:25
i have solve the issue by changing the sensor.
to rate and post a comment
Bao Vo Quoc
re: re: re: AGFA Compact Plus Error 22862 on selftest
October 06, 2012 12:52
in my opinion
The problem is causinf from 3 factor as below:
1. Control PCB for Rotation is failed
2. Light Barrier sensor of Control PCB Rotation failed.
3. Light Barrier of A Cassette module is failed
try to clean all these above sensor and check again
Good Luck to you
Bao
to rate and post a comment
Jose Garcia
re: re: AGFA Compact Plus Error 22862 on selftest
October 03, 2012 06:15
Bao:DoYOu have service manual of this equipment
thanks
Jose L.Garcia
to rate and post a comment
TOM CHARLES
re: re: re: AGFA Compact Plus Error 22862 on selftest
October 10, 2012 02:30
Jose email me at tomcharles41@yahoo.com
to rate and post a comment