ADC SOLO 5155/100 ERROR CODE 21012
November 07, 2019 09:09
Hello there
I have model Agfa Adc solo Cr 5155/100 device.
as error code : 21012. (Reason:cannot create control file. Creation of the file containing the control data failed. )
I checked the HDD
there was a bad sector
I found a solid hdd
I made one to one copy
I did hdd repair through the system
Again gives the same error
Do you know anything about this topic
Thanks
to rate and post a comment
Ronald Swapp
re: ADC SOLO 5155/100 ERROR CODE 21012
November 08, 2019 10:38
I'm assuming you did a one to one copy from a scsi to a scsi drive. Also you have to use the HD Formatter tool which should be on a floppy. You need to monitor via hyperterm or Putty and have a d9 to d25 serial cable to do this properly.