Ray Gonzalez

Hologic QDR 1000 Plus

October 05, 2004 08:35

Need help with a Hologic QDR 1000 Plus. I am getting the error ("Timeout: no 120Hz interrupts. This points out to the Timing and Control PCB.I have replaced all the boards on the unit including the computer and still get the same error.
Any help will be apreciatted.
Thanks in advance.
Please reply to: Ray Gonzalez
ray@4biomed.com

Log inor Register

to rate and post a comment

(128)
(10)
(5)

Manuel Esquer

Re: Hologic QDR 1000 Plus

October 13, 2004 02:32

Verify chopper pickoff sincronization
Check filter wheel pickoff
Check timing & control board and reseat

Log inor Register

to rate and post a comment

(18)
(2)
(2)

Ray Gonzalez

Re: Hologic QDR 1000 Plus

October 13, 2004 07:05

We appreciate your help Manuel. The problem was the x-ray tank. We just replaced it thanks to the fact that we have one unit for parts.

Regards,
Ray Gonzalez

Log inor Register