I've a GE Lunar DPX that consistently have a QA failure, in the spillover test. It gets more than 12%. But, sometimes, if we insist more than 3-4 times, the spillover get under 12% (11,97%) and the QA test will pass.
There's some thing i can do, and can someone send me the service manual, also?
re: GE Lunar DPX (Pro) QA failure - service manual
November 04, 2014 05:37
You likely have a bad detector. You can sometimes get past this by using the x-ray test mode to saturate the detector. This is what happens when you run the QA multiple times back to back. The detector is easy to replace but you will likely have to run the peak alignment. Just remember that the detector is not properly responsive during QC and exams as well. You may be getting improper results on the patient.
re: GE Lunar DPX (Pro) QA failure - service manual
November 05, 2014 08:08
really what the friend said Desmond Johnson really is the truth, follow the link to the manula, sorry to be in Portuguese.
https://drive.google.com/file/d/0B7EG7VLkEySEYUlELWZOMGNleUk/view?usp=sharing
re: GE Lunar DPX (Pro) QA failure - service manual
November 05, 2014 09:58
If you like i find you a reloaded detector. you can also connect the system with teamviewer to help you for the installation and/or give your detector one more year life.
service@alfatech.gr
re: GE Lunar DPX (Pro) QA failure - service manual
November 05, 2014 10:00
Hi Adriano,
As mentioned, if nothing has changed recently with alignment, then your detector has deteriorated. This is a very common failure on any of the DPX models. Having worked for the Lunar business for many years we have seen these failures numerous times. We too stock these detectors. We have both new and used options to fit your budget. Also, as was mentioned, this is a part you do not need any specialty tools to replace if the alignment has not changed. Additionally, we season the detectors in a system prior to shipping to ensure quality and to save you from having to complete that process. If you would like a quote feel free to email me at ben@dexasolutions.com.
re: re: GE Lunar DPX (Pro) QA failure - service manual
September 27, 2015 02:51
Hi Ben, I would like get you help.
I am checking a DPX IQ and have 2 problems:
1.- The accuracy in the spine fantom is dropping. On the momento the bmd is in 1,190 (L2-L4,normal ranging 1,245 to 1,311), the last reading before of the fail was 1,240.
2.- The collimation ratio is 3.41 (Fail)
Other meassure:
Spillopver is 9,2;}
Low count: 530000 and hi count:316000 approx.
Tissue value 1,308
re: re: GE Lunar DPX (Pro) QA failure - service manual
October 06, 2022 09:35
Hello all I'm having the same issue with the spectrum spillover is 14.22% error the status which is in failed any suggestions to up the error. Between how to do the peak alignment to check?
re: GE Lunar DPX (Pro) QA failure - service manual
December 15, 2016 02:59
Had the same problem was bad detector . sometimes you can open the detector and clean the windows sometimes they tarnish and cleaning will fix the saturation problem.
Altima will inspire the trust, loyalty and commitment of our customers through positive relationships and collaboration, top quality equipment services, and a clearly superior value proposition. Call Today (469) 644-7063
Adriano Araújo
GE Lunar DPX (Pro) QA failure - service manual
November 04, 2014 04:23
Greetings,
I've a GE Lunar DPX that consistently have a QA failure, in the spillover test. It gets more than 12%. But, sometimes, if we insist more than 3-4 times, the spillover get under 12% (11,97%) and the QA test will pass.
There's some thing i can do, and can someone send me the service manual, also?
Thank you all.
to rate and post a comment
Desmond Johnson
re: GE Lunar DPX (Pro) QA failure - service manual
November 04, 2014 05:37
You likely have a bad detector. You can sometimes get past this by using the x-ray test mode to saturate the detector. This is what happens when you run the QA multiple times back to back. The detector is easy to replace but you will likely have to run the peak alignment. Just remember that the detector is not properly responsive during QC and exams as well. You may be getting improper results on the patient.
to rate and post a comment
LOUIE
re: GE Lunar DPX (Pro) QA failure - service manual
November 05, 2014 08:08
You probably have a bad detector if the home positioning is good. We have it in stock if you need one
to rate and post a comment
Lenny Place
re: GE Lunar DPX (Pro) QA failure - service manual
November 05, 2014 08:08
send me an email address. this could be the detector but I would go thru the alignments and calibrations first.
to rate and post a comment
ANDRE FREIRE
re: GE Lunar DPX (Pro) QA failure - service manual
November 05, 2014 08:08
really what the friend said Desmond Johnson really is the truth, follow the link to the manula, sorry to be in Portuguese.
https://drive.google.com/file/d/0B7EG7VLkEySEYUlELWZOMGNleUk/view?usp=sharing
to rate and post a comment
Adriano Araújo
re: re: GE Lunar DPX (Pro) QA failure - service manual
November 05, 2014 09:59
Thank you. The language is no problem, actually is perfect. I'm Portuguese :)
to rate and post a comment
Carlos Carvalho
re: re: re: GE Lunar DPX (Pro) QA failure - service manual
December 15, 2016 10:20
Hello,
Can you send me the manual of GE Lunar DPX because the link isn't working anymore. Thank you
to rate and post a comment
hossam el farghaly
re: re: GE Lunar DPX (Pro) QA failure - service manual
November 08, 2014 09:32
Hi André, do u have the service manual of Prodigy but in English? Thanks
to rate and post a comment
Carlos Carvalho
re: re: GE Lunar DPX (Pro) QA failure - service manual
December 15, 2016 10:20
The link isn't working. Could you send me the manual to me? Thank you
to rate and post a comment
CHRIS NALMPANTIS
re: GE Lunar DPX (Pro) QA failure - service manual
November 05, 2014 09:58
If you like i find you a reloaded detector. you can also connect the system with teamviewer to help you for the installation and/or give your detector one more year life.
service@alfatech.gr
to rate and post a comment
Ben Williams
re: GE Lunar DPX (Pro) QA failure - service manual
November 05, 2014 10:00
Hi Adriano,
As mentioned, if nothing has changed recently with alignment, then your detector has deteriorated. This is a very common failure on any of the DPX models. Having worked for the Lunar business for many years we have seen these failures numerous times. We too stock these detectors. We have both new and used options to fit your budget. Also, as was mentioned, this is a part you do not need any specialty tools to replace if the alignment has not changed. Additionally, we season the detectors in a system prior to shipping to ensure quality and to save you from having to complete that process. If you would like a quote feel free to email me at ben@dexasolutions.com.
Thanks,
Ben
to rate and post a comment
mario martinez
re: re: GE Lunar DPX (Pro) QA failure - service manual
September 27, 2015 02:51
Hi Ben, I would like get you help.
I am checking a DPX IQ and have 2 problems:
1.- The accuracy in the spine fantom is dropping. On the momento the bmd is in 1,190 (L2-L4,normal ranging 1,245 to 1,311), the last reading before of the fail was 1,240.
2.- The collimation ratio is 3.41 (Fail)
Other meassure:
Spillopver is 9,2;}
Low count: 530000 and hi count:316000 approx.
Tissue value 1,308
Waiting for a prompt answer.
Best refards.
Mario
to rate and post a comment
Mohamed Ashik
re: re: GE Lunar DPX (Pro) QA failure - service manual
October 06, 2022 09:35
Hello all I'm having the same issue with the spectrum spillover is 14.22% error the status which is in failed any suggestions to up the error. Between how to do the peak alignment to check?
to rate and post a comment
Ben Williams
re: re: re: GE Lunar DPX (Pro) QA failure - service manual
October 06, 2022 01:10
You will most likely need a detector unless the alignment is off. The alignment should be done by taking a picture.
to rate and post a comment
Mike Hopkins
re: GE Lunar DPX (Pro) QA failure - service manual
December 15, 2016 02:59
Had the same problem was bad detector . sometimes you can open the detector and clean the windows sometimes they tarnish and cleaning will fix the saturation problem.