Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

GE lightspeed 16 reconstruction failure.

Tags: CT, GE HEALTHCARE

(2)

Michael Nerubayskiy

GE lightspeed 16 reconstruction failure.

February 22, 2024 12:27

Hi everyone. I have lightspeed 16 with DARC Jarrel and 1 image generator.
Some time ago it failed to do recon and scsi raw data drives failed.
So I changed hard drives in DARC and now it’s working but ct doesn’t reconstruct images anyway. So I replaced RAC board because it’s notorious of malfunctions but it doesn’t changed the situation (RAC board is used and seller claims it’s in operational condition).
I plugged monitor to Image generator and what i found out is: it starts to deploy os from DARC and stucks at “Uncrompressing Linux… ok, booting the kernel”. Nothing happens after. Systems couple times tries to reset IG and then nothing.
I already reinstalled DARC OS using Linux shell on the console , but nothing has changed. Same issues.
Please help,

Log inor Register

to rate and post a comment

(6)

shams tabrej

re: GE lightspeed 16 reconstruction failure.

February 23, 2024 10:52

Dear. Sir in darc bios once u cheq Darc processor. Both processor are showing or not..and second change Darc ram also ..if ok bypass ig by command

Log inor Register

to rate and post a comment

(38)
(2)
(4)

Claude Coelo

re: re: GE lightspeed 16 reconstruction failure.

February 23, 2024 04:03

Hello
What is the software version ?

Log inor Register

to rate and post a comment

(46)
(3)

Carlos Pons

re: re: re: GE lightspeed 16 reconstruction failure.

February 26, 2024 01:26

check error log on darc, check ping ig1 from darc ,should be ok, send or attach logs, seems your darc is not comunicating with IG

Log inor Register

to rate and post a comment

(15)

(1)

Arthur Singh

re: GE lightspeed 16 reconstruction failure.

February 29, 2024 09:56

Hi. when you replace scsi hdd, did you check jumpers? you have to make same scsi id like id0, id1 like it was before. check in darc bios setting if scsi hdd is showing or not? if raw data hdd are not present darc may not boot properly. after detecting of raw data disks, please reinstall darc software. good luck.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Próximos eventos de leilão

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)