Hi, dotmed stuff!
The device starts - fine. But when the X-ray process begins, the following error appears on the display:
"Fault
Measured kV UNDER"
Photo: https://ibb.co/6Pn1GQG
Nowhere about this error is not written. Board checking everything is fine.
Tell me please exactly what this error is connected with?

Log inor Register

to rate and post a comment

(48)
(4)

Emrah Kambak

re:Shimadzu UD-150L-40E Fault Measured kV UNDER ERROR

March 15, 2019 08:06

Hi. My name is emrah. Please check inverter fuse. 100A. If blown check igbt.

Log inor Register

to rate and post a comment

(2)

Insendio fx

re: re:Shimadzu UD-150L-40E Fault Measured kV UNDER ERROR

March 15, 2019 10:59

Hi, Emrah. We have already checked the inverter fuse - no defective inverters were found. Tomorrow check igbt. Thanks for your reply.

Log inor Register

to rate and post a comment

(38)
(3)
(2)

Alessandro Lazzari

re: Shimadzu UD-150L-40E Fault Measured kV UNDER ERROR

March 15, 2019 08:06

This error means that kv feedback is

lower then expected. Measure by
oscilloscope kvfb+ and - and compare
with the set output valute to evacuate where is the fault

Log inor Register

to rate and post a comment

(48)
(4)

Emrah Kambak

re:Shimadzu UD-150L-40E Fault Measured kV UNDER ERROR

March 15, 2019 08:07

Please contact me ekambak@gmail.com

Log inor Register

to rate and post a comment

(38)
(3)
(2)

Alessandro Lazzari

re: Shimadzu UD-150L-40E Fault Measured kV UNDER ERROR

March 15, 2019 08:08

Hi ! I see this error before. You have to check kv+/- feedback that comes from HV tank and compare to real Kv setting measured then you can judge and try to adjust the lower value or continue troubleshooting on Hv control-Hv-Tube

Log inor Register

to rate and post a comment

Complete Medical Services - Leader in New & Pre-Owned Bone Densitometers

NEW and Gold Seal GE Lunar Bone Densitometers (US Only). GE Lunar Prodigy, GE Lunar iDXA. Pre-Owned available worldwide. Contact Us Now - 586-532-1142 DOTmed Certified

 

advertisement

 

 

advertisement

 

advertisement
Advertisement
O acesso e o uso deste local são sujeitos aos termos e às condições do nosso OBSERVAÇÃO LEGAL & OBSERVAÇÃO DA PRIVACIDADE
Propriedade de e proprietário DOTmeda .com, Inc. Copyright ©2001-2019 DOTmed.com, Inc.
TODOS OS DIREITOS RESERVADOS