Kodak 8000 Error - Preste serviços de manutenção a técnicos Forum
advertisement

Kodak 8000 error

Tags: Dental, KODAK

(5)
(3)

Svyatoslav Timchenko

Kodak 8000 error

July 21, 2016 09:59

Hi, guys
I've got Kodak 8000 PCi-board version. Occasionally, but quite often it gives errors 201, 410 or 416. I tried to reinstall software on another pc - it helps for a while, then - same problems with image reception. Any solution? Also, could you send me a copy of service manual.
Thank you!

1 person liked this
Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: Kodak 8000 error

July 22, 2016 10:34

The service manuals don't have those errors you mention. You must have the addendum service manual fot the description. These two have been sent to you by mail today.
Regards

3 people liked this
Log inor Register

to rate and post a comment

(5)
(3)

Svyatoslav Timchenko

re: re: Kodak 8000 error

July 22, 2016 02:47

thank you very much , Faustino, for sending the manuals to me. I managed to find those errors, that i mentioned. They appear on the computer monitor:
error 201 - Problems with image reception
Possible cause:
Low voltage power supply to the digital sensor is defective.
Remarks:
The sensor can also be defective
Identification of the default:
No image transfer.
Actions:
Replace the power supply and check with another sensor

Log inor Register

to rate and post a comment

(5)
(3)

Svyatoslav Timchenko

re: re: re: Kodak 8000 error

July 22, 2016 03:39

E410 The panoramic board failed to acquire an image
Possible cause:
Delay too long between the beginning of the exposure and the reception of the first pixels.
Remarks:
This error might appear after several E 201.

Actions:
Refer to E 201 Possible cause:
Delay too long between the beginning of the exposure and the reception of the first pixels.
Remarks:
This error might appear after several E 201.
Identification of the default:
Actions:Refer to E 201
There is no E416 in the service manual.

On the unit the last error displayed was e22

I wonder, is there a way to check if it is a defect of sensor without replacing it?

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: Kodak 8000 error

July 22, 2016 05:07

Err 22 and 201 are the same, most of them are due to beam misaligment of bad power supply for the sensor (black box). As the unit has certainly done a lot of exposures, the cables on the rotation could be defective, mainly the wires of the image or those from the encoder. Send me an email and we could arrange some way of sending you a sensor and a power supply for testing.

Log inor Register

to rate and post a comment

(5)
(3)

Svyatoslav Timchenko

re: re: re: re: re: Kodak 8000 error

July 23, 2016 12:21

my e-mail is slava500@gmail.com
Will try to check the cable and voltage first.

Log inor Register

to rate and post a comment

(5)
(3)

Svyatoslav Timchenko

re: re: re: re: re: re: Kodak 8000 error

July 26, 2016 10:04

Managed to fix it!!! That was a faulty condenser in the power supply to digital sensor. Feel happy. Thanks again, Faustino.

2 people liked this
Log inor Register

to rate and post a comment

(2)

sonia escobar

re: re: re: re: re: Kodak 8000 error

October 09, 2018 10:24

Hi my kodak 8000c is also giving me erroe code panoramic board fail to acquire image

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: Kodak 8000 error

October 10, 2018 08:46

Link to access tech files sent by Email Oct 9. Gracias

Log inor Register

to rate and post a comment

(2)

GERMAN MENDEZ

re: re: re: re: re: re: re: Kodak 8000 error

September 06, 2019 09:28

Hi Mr Faustino , i have a Trophy Pan 8000 , that has an error E53 , i would like to know what part should i buy to change and where can i buy it

Log inor Register

to rate and post a comment

(3)

Nguyen Nguyen

re: re: re: re: re: Kodak 8000 error

June 21, 2019 08:00

We have Kodak 8000 PCI version. The rotating arm does not complete the rotation and Error n*201 "Problem with image reception" appears. Also, Error #410 "The panoramic board failed to acquire an image" follows afterward. What could be the problems? Is Error n*201 same as error E 201 in the Addendum? By the way I use * in place of what appear to be a degree (o) symbol in Error n*201. Thank you!

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: Kodak 8000 error

June 24, 2019 10:28

The answer is also on the top of this discussion by Svyatoslav Timchenko. You might have the same problem.

Log inor Register

to rate and post a comment

(10)

Vicent Martinez

re: re: Kodak 8000 error

November 17, 2017 09:22

Hello Faustino. Can you send me the service manual of the kodak 8000. I have err 53. My email is vgodayol@electromedical.net

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: Kodak 8000 error

November 17, 2017 05:46

Ok, You will receive a link by email, please "like" my comments, it will make me happy.
Regards to all readers

5 people liked this
Log inor Register

to rate and post a comment

(1)
(1)

Joao Branco

re: re: re: re: Kodak 8000 error

May 20, 2018 03:19

Hello Faustino. Can you send me the service manual. I have error E53 on my Kodak 8000. My email is joaoduque1974@gmail.com

1 person liked this
Log inor Register

to rate and post a comment

(3)

Kornel Csongrady

re: re: re: re: re: Kodak 8000 error

February 27, 2019 10:39

Hi, I have a same massage. Do you have the addendum service manual? can you send it tome? Thanks.

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: Kodak 8000 error

February 28, 2019 09:46

Information sent by dotmed email service.Thanks

Log inor Register

to rate and post a comment

(1)

Simon Locke

re: re: re: re: re: re: re: Kodak 8000 error

July 19, 2019 10:20

Hi
I have an error 53 on my Kodak 8000 too. Would you be so kind as to send me the information to fix this please?

Thanks
Simon

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: re: re: Kodak 8000 error

July 19, 2019 10:46

Info sent by Dotmed mail service

Log inor Register

to rate and post a comment

(1)

Daniel Martins

re: re: re: re: Kodak 8000 error

July 06, 2018 10:36

Hi Faustino, I have a error E416 . Can you help me?

Regards
Daniel

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: Kodak 8000 error

August 01, 2018 08:58

I think I have sent you a technical manual

Log inor Register

to rate and post a comment

Alexandr Mochalov

re: re: re: re: Kodak 8000 error

July 31, 2018 10:30

Good day! Help please in solving the problem with the error e53!

Log inor Register

to rate and post a comment

(2)

sonia escobar

re: re: re: re: re: Kodak 8000 error

October 09, 2018 10:24

Hi i need help as well my kodak 8000c is giving me error message stating panoramic board failed to acquire image

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: Kodak 8000 error

October 10, 2018 08:46

Already sent link to tech files by mail, hope you can manage to repair following my instructions, Thanks for reading.

1 person liked this
Log inor Register

to rate and post a comment

(13)

migel elizarraras

re: re: re: re: re: re: re: Kodak 8000 error

October 23, 2018 09:03

Good afternoon Faustino:

If you have the service manual of kodack 8000C, can you sent me a copy to my email?

Thank a lot

miguelelizarrarasm@hotmail.com

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: re: re: Kodak 8000 error

October 23, 2018 04:46

Sent by mail today 23/10

1 person liked this
Log inor Register

to rate and post a comment

(1)

Skender Ahmetaj

re: re: re: re: Kodak 8000 error

July 16, 2019 11:28

Dear Faustino,
I have Kodac8000C XRay machine. The error code 53 is coming up with a low backup battery. Do you know where the battery is located in the Xray machine? Could you please send me some information about it? Many thanks.

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: Kodak 8000 error

July 17, 2019 09:34

information sent to you trough Dotmed message center,
regards

Log inor Register

to rate and post a comment

(3)

ALFATECH

re: re: Kodak 8000 error

December 01, 2018 01:21

Can you pls send me the service manual? I have an error E53. Thank you in advance. My email is info@alfatech.gr

Log inor Register

to rate and post a comment

(2)

Kat Almond

re: re: Kodak 8000 error

December 18, 2018 09:46

kalmond@aipdbr.com error E53
Thank you!

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: Kodak 8000 error

December 19, 2018 09:09

sent by mail

1 person liked this
Log inor Register

to rate and post a comment

(3)

Kornel Csongrady

re: re: Kodak 8000 error

February 27, 2019 10:39

Dear Faustino,
I have Kodac8000C XRay machine. The error code 53 is coming up with a low backup battery. Do you know where the battery is located in the Xray machine? Could you please send me some information about it? Many thanks.

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: Kodak 8000 error

February 28, 2019 09:46

Information sent by Dotmed mail service, Thanks

Log inor Register

to rate and post a comment

(1)

tesha whaley

re: re: re: re: Kodak 8000 error

June 12, 2019 10:02

I'm having the same error this morning as Kornel did with a low backup battery error that won't let me take an xray. Can you help me?

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: Kodak 8000 error

June 13, 2019 10:19

Information has been sent by Dotmed message center. Hope it helps. In case of non receipt, just email me. Thanks

Log inor Register

to rate and post a comment

(2)

GERMAN MENDEZ

re: re: Kodak 8000 error

September 06, 2019 09:24

HI MR FAUSTINO i own a trophy pan 8000 that has a error E53 , i would like to know what part should i buy and change ... thanks

Log inor Register

to rate and post a comment

(2)

Kat Almond

re: Kodak 8000 error

December 18, 2018 09:46

I also need the Kodak 8000c addendum service manual- error E53. Thank you!

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: Kodak 8000 error

December 19, 2018 09:09

This error is due to the RTC battery that is dead. Link has advise how to replace battery on M48 Ic's.

1 person liked this
Log inor Register

to rate and post a comment

Elton Gjoshi

re: re: re: Kodak 8000 error

December 26, 2018 08:49

Hi Faustino!I have a problem with my kodak 8000 error#311 please can you help me. thankyou

Log inor Register

to rate and post a comment

(10)

MatasRoldan

re: re: re: re: Kodak 8000 error

December 26, 2018 05:07

gime your mail or write me danielrkoorton@gmail.com

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: Kodak 8000 error

December 26, 2018 05:09

Hi Elton, there are 2 K8K models; one is a PCI version, the other an Ethernet version. For both models the 311 error concerns the digital sensor not being detected. For the PCI version the cause might be the black power supply, cables (from sensor to PC), the sensor, the PCI board, RS422 cable was extended, etc. Link to access service manuals has been sent trough Dotmed message center as you didn't mention any email; Thanks for reading....

Log inor Register

to rate and post a comment

(2)
(1)

Joseph Dixon

re: re: re: Kodak 8000 error

February 26, 2019 10:18

Faustino would you resend a link to me on how to replace the battery I have an 8000 with the error E53 I would greatly appreciate it Joe Dixon avatardentalsolutions@gmail.com

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: Kodak 8000 error

February 27, 2019 10:38

Information has been sent to you email, Regards

Log inor Register

to rate and post a comment

(3)
(1)

Octavian Sarca

re: re: re: re: re: Kodak 8000 error

March 31, 2019 04:10

I have the same problem here and I woudl greatly appreciate your help. My email is : tavisarca@gmail.com

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: Kodak 8000 error

April 01, 2019 09:11

Info sent to your mail.

Log inor Register

to rate and post a comment

(3)
(1)

Octavian Sarca

re: re: re: re: re: re: re: Kodak 8000 error

May 26, 2019 12:02

Dear Faustino, thank you very much for all the help. I received and installed the components from Amazon, I got E58, but disappeared after the fist reboot. This is a very small dental office, with limited budget. You helped us save thousands of dollars. We wish you a long and healthy life!!!

1 person liked this
Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: re: re: Kodak 8000 error

May 27, 2019 10:46

Thanks for your nice comment, glad that it works.

Log inor Register

to rate and post a comment

(3)

Kornel Csongrady

re: re: re: re: Kodak 8000 error

February 27, 2019 10:39

Hi, I have the same problem. Could you please send me some information how to replace the backup battery on Kodac8000C? Many thanks.

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: Kodak 8000 error

February 28, 2019 09:46

Information sent by Dotmed mail service as you didn't give any email.
Rgds

Log inor Register

to rate and post a comment

(1)

jean didelot

re: re: re: re: re: re: Kodak 8000 error

March 20, 2019 09:57

Dear, I have E53 error and also E58 after replacing the chip, and I need the service manual. It would be wonderful if you could send it to me! My email is jm.didelot@teraneo.fr
Thanks !

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: re: Kodak 8000 error

March 21, 2019 09:43

Information envoyée par email, Merci

Log inor Register

to rate and post a comment

(3)
(1)

Octavian Sarca

re: re: re: Kodak 8000 error

March 31, 2019 04:10

Hello Faustino, I have a Kodak 8000 (PCI version) with error E53. Could you please send me the service manual, the addendum and any other instructions to help me replace the battery. Thank you!!!

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: Kodak 8000 error

April 01, 2019 09:11

Info already sent to your mail.

Log inor Register

to rate and post a comment

(3)

Mahdi Ali

re: re: re: re: re: Kodak 8000 error

April 29, 2019 09:29

hello

could you send me instructions on how to fix E58 in kodak. I replaced the battery on board cj644 the PC communication board and now this error is coming up. let mek now if you could help.

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: Kodak 8000 error

April 30, 2019 10:21

Hello, If the battery is good, then reset the error and take a picture,On the next power ON, you will not get the error again.

Log inor Register

to rate and post a comment

(3)

Mahdi Ali

re: re: re: re: re: re: re: Kodak 8000 error

May 13, 2019 10:32

how do you reset the error? I turned the machine off and on and the error still comes back. is there a certain way to push the buttons to reset the machine or is there a button to press to rest. please advise. and if you could send me the manual.

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: re: re: Kodak 8000 error

May 14, 2019 09:47

info sent to you by mail

Log inor Register

to rate and post a comment

(3)

Mahdi Ali

re: re: re: re: re: re: re: Kodak 8000 error

May 12, 2019 12:41

my email is

drmahdiali@yahoo.com

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: re: re: Kodak 8000 error

May 14, 2019 09:47

OK

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: Kodak 8000 error

May 14, 2019 09:47

info sent by email

Log inor Register

to rate and post a comment

(2)

don pierce

re: re: re: re: re: Kodak 8000 error

May 10, 2019 10:50

may i also have the service manual? don.pierce@infinitepc.net

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: Kodak 8000 error

May 13, 2019 10:37

Link sent by mail for the K8000 unit.
The 8100 is different and I don't have them, Regards

Log inor Register

to rate and post a comment

(3)

Nguyen Nguyen

re: re: re: re: re: Kodak 8000 error

June 11, 2019 08:55

Could you please send me the service manual, the addendum as well? Thank you very much!!!

Log inor Register

to rate and post a comment

(2)

don pierce

re: Kodak 8000 error

May 10, 2019 10:50

Hello Faustino. Can you send me the service manual of the kodak 8100 I have err 21. My email is don.pierce@infinitepc.net

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: Kodak 8000 error

May 13, 2019 10:37

The 8100 is completely different and I don't have those manuals. Regards

Log inor Register

to rate and post a comment

(1)

Mark Robertson

re: re: re: Kodak 8000 error

June 04, 2019 08:17

I'm having the e53 error code as well could you send the manual to markrobertsondds@gmail.com thanks

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: Kodak 8000 error

June 05, 2019 08:47

Hi, information has been sent to your Email. Glad to have been of some help. Bye

Log inor Register

to rate and post a comment

(3)

Nguyen Nguyen

re: re: re: re: re: Kodak 8000 error

June 11, 2019 08:54

Hello Faustino. We have Kodak 8000 with PCI. Can you send me the service manual? My email is housemls@yahoo.com. Thank you!

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: re: Kodak 8000 error

June 12, 2019 10:01

Link to tech docs sent by email, Thanks

Log inor Register

to rate and post a comment

(4)

Roberto Rustici

re: re: re: re: Kodak 8000 error

July 04, 2019 10:06

Hi, Faustino, same E53 error, may I have the service manual?
Thankyou in advance
roberto@multimedicava.com

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: re: Kodak 8000 error

July 05, 2019 11:35

Link to tech docs has been sent to your email.
Thanks

Log inor Register

to rate and post a comment

(30)
(3)

TUTUL

re: Kodak 8000 error

July 05, 2019 02:52

i need GE BRIVO XR 115 SERVICE MANUAL.
pls send me a copy .partha.saha036@gmail.com

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: Kodak 8000 error

July 05, 2019 05:25

Sorry but I don't have anything about GE. This topic is about Kodak 8000 panoramic unit

Log inor Register

to rate and post a comment

(1)

Zaru

re: Kodak 8000 error

July 08, 2019 12:10

Hi
I also have the same problem E53 error and low battery error.
Kindly forward me the service manual details. Thank you.
zarushin@yahoo.com

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: Kodak 8000 error

July 09, 2019 10:24

Info sent to you by Dotmed message center, Tks

Log inor Register

to rate and post a comment

(2)

Stefan Wenk

re: re: re: Kodak 8000 error

July 31, 2019 04:32

Hi,
I also have the E53 problem, can you send me the manual to fix it?
thx

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: re: re: Kodak 8000 error

August 01, 2019 10:25

Info sent by email, thanks

1 person liked this
Log inor Register

to rate and post a comment

(2)

Stefan Wenk

re: re: re: re: re: Kodak 8000 error

August 02, 2019 09:30

Thanks for your help. My Kodak 8000 is repaired!

Log inor Register

to rate and post a comment

(295)
(119)
(2)

Faustino Pereira

re: re: Kodak 8000 error

July 09, 2019 10:25

Link sent to your email, thanks

Log inor Register

to rate and post a comment

(1)

aron warner

re: re: re: Kodak 8000 error

August 04, 2019 12:08

The solution you are providing is proper. I want to add something with this. If you want to get any kind of solution regarding the router then you can go through [url=https://netgears.support/blog/solve-netgear-router-password-issues/]Solve Netgear Router Password Issues[/url] for the best solution.

Log inor Register

to rate and post a comment

SERVICE ALL C-arms - SPECIALIZE in OEC 9800 & 9900 | CALL CMS 586-532-1142

National service coverage with a well trained and professional team. Complete Medical Services has over 20 years of experience in this industry specializing in OEC C-arm equipment. LOOKING for a pre-owned C-arm - CALL we always have systems in in-stock

advertisement

 

advertisement

 

advertisement

 

advertisement