Innova 2000 Angiography - Preste serviços de manutenção a técnicos Forum

Innova 2000 angiography

Tags: Cardiac Imaging

Denis Kim

Innova 2000 angiography

January 27, 2020 10:23

Hello. I have some problems with innova GE 2000 angiography. Tube moving is very slow sometimes. I don’t know what the problem is. Can some body help me with this? Thank you)))

Log inor Register

to rate and post a comment

(75)
(7)

R GOPAL

re: Innova 2000 angiography

January 28, 2020 12:18

You mean anode rotation?

Log inor Register

to rate and post a comment

(2)

Larry James

re: Innova 2000 angiography

February 13, 2020 09:27

I assume you are having issues with the C-arm moving slow. This issue is a difficult problem to address. Can you tell me about the errors in the PST and the error list on the advantx. Have you looked at the bump device around the detector? There is 1st level and 2nd level. You can have issues with the air device or the micro switch's around the detector

Log inor Register

to rate and post a comment