dismiss

Clean Sweep Live Auction on Wed. February 27th. Click to view the full inventory

SEARCH
Posição atual:
>
>
>
> I NEED SERVICE MANUAL Polydoros LX 50

I NEED SERVICE MANUAL Polydoros LX 50

Tags: Radiology - R/F, SIEMENS

(17)
(1)

MONA

I NEED SERVICE MANUAL Polydoros LX 50

January 29, 2014 08:24

Dear All
I HAVE ERROR 606
Looking for a service manual for Routine X-Ray, Polydoros LX 50 SIEMENS.

Regards

Log inor Register

to rate and post a comment

(2596)
(549)
(15)

Lenny Place

re: I NEED SERVICE MANUAL Polydoros LX 50

January 29, 2014 06:13

I can send it to you email address please
606 Meaning:
RANGE Error


Description:
The signals RANGE0 ”OK” and RANGE1 ”OK” monitor the switchover of the tube current measuring ranges on D220. The corresponding relays are monitored via the RANGE 0, and RANGE 1 signals.


Possible causes and action:
Check function at SS OFF on D220. Test points: D220.X19.A1, A2
(wiring diagram X2206-18 and 22 / X2169-18 and 22):

1. With ZB ON: measuring range 200mA;
"RANGE 0" X19.A1 = 0V -> "RANGE 0" ”OK” X19.A4 = 24V (±15%)

"RANGE 1" without significance

2. With normal FL ON: measuring range 1mA;
"RANGE 0" X19.A1 = 24 V (±15%)
"RANGE 0" ”OK” X19.A4 = 0V
RANGE 1" X19.A2 = 24V (±15%) -> "RANGE 1" ”OK” X19.A3 = 0V

If the "RANGE 0" and "RANGE 1" signals do not agree with the operating mode, there is a fault in the cable X1 to D100 or on D100.

If the signals "RANGE 0"_OK and "RANGE 1"_OK are wrong with correct actuation, D220 must be replaced.




XCS ©2004 by Siemens AG Med AXPLM-T SC

Log inor Register

to rate and post a comment

(17)
(1)

MONA

re: re: I NEED SERVICE MANUAL Polydoros LX 50

January 31, 2014 04:25

Dear Lenny Place

Thank you for your guidance.
Could you please send for me service manual ?
my email : monataqavi@gmail.com

thank you & Best regards
Mona

Log inor Register

to rate and post a comment

(190)
(17)

Luis Fernando Garcia Sanchez

re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

December 02, 2015 09:43

Could you send me?, service manual polydoros lx. Email luisfernandogarcias@yahoo.com regards

Log inor Register

to rate and post a comment

(506)
(43)

DMEDS

re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

December 09, 2015 08:50

i have all manuals for this system

PolyLX_Adjustment
PolyLX_Troubleshooting
PolyLX_Wiring and schematic diagram
Replacement of Parts

contact me

magmesme@gmail.com

regards

Log inor Register

to rate and post a comment

(10)

Alejandro Orihuela

re: re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

February 06, 2019 08:48

regards!! I need the manual of polydoros lx 50, I get error 805 and 777 thanks
Please send me the service manual please
alejandro_orihuela@hotmail.com

Log inor Register

to rate and post a comment

(2596)
(549)
(15)

Lenny Place

re: re: re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

February 06, 2019 09:10

Look for an email from a site called myairbridge

Log inor Register

to rate and post a comment

(203)
(36)

Amos Dantene

re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

November 07, 2018 08:21

Please I need the service manual. My email address is: adantene@gmail.com

Log inor Register

to rate and post a comment

(2596)
(549)
(15)

Lenny Place

re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

November 08, 2018 09:55

Look for an email from a site called myairbridge

Log inor Register

to rate and post a comment

(2)

Rama Kant

re: re: I NEED SERVICE MANUAL Polydoros LX 50

June 12, 2015 03:44

Dear Lenny Place,

I shall be thankful for sending me servicr manual of Polydoros LX 50 at my mail reconsystems@yahoo.com

Regards

Rama Kant

Log inor Register

to rate and post a comment

(506)
(43)

DMEDS

re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

December 09, 2015 08:50

i have all manuals for this system

PolyLX_Adjustment
PolyLX_Troubleshooting
PolyLX_Wiring and schematic diagram
Replacement of Parts

contact me

magmesme@gmail.com

regards

Log inor Register

to rate and post a comment

(13)
(1)

nos

re: re: I NEED SERVICE MANUAL Polydoros LX 50

April 17, 2018 10:47

what does err 119/080 for generator POLYDOROS LX 50 mean

Log inor Register

to rate and post a comment

(2596)
(549)
(15)

Lenny Place

re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

April 17, 2018 05:58

Error 119

Designation:
Heating circuit timeout
Description:
Communication between the D220 heating circuit and the D100 master takes place over the serial interface (MPS, D100.X1.28, D220.X1.28)
If an interrupt of communication between the D100 and D220 takes place, error 119 is sent out by the D100.
Possible Causes of Error and Remedies:
· Serial interface interrupted, check →
· D100, D220 defective, replace →
· Arcing tube
· H1 high voltage transformer
· Check how the cables are laid (shielding)
· Check the ground wires, all ground wire connections provided by the manufacturer must be connected.
· Check the firmware version.
· Check the 5V power supply on the SNT, D160 and D220.
- Measure at the SNT X2.6 (5V) switch power supply versus X2.3 (0V), set approx. 5.1V (max. 5.2V).
- On the D160. Measure X31 (5V) versus X28, the voltage drop versus the SNT should be less than 20mV.
- Measure on D220.X18.A1 (5V) versus X15, the voltage drop versus the SNT should be less than 150mV.
- If the voltage drop on the D220 is more than 150mV, check the following plug-in connections: SNT.X2, D160.X11, D160.X13, D100.X13, D100.X1, D220.X1.
- If the X1 ribbon cable between D100 and D2209 is causing the voltage drop, replace it with a new one with Part No. 7463677.

· Observe the generator Overview Wiring Diagram for lamps and LEDs in the power cabinet.
· Ignore error 119 if the red LED (D220) is blinking and no firmware is loaded; perform a → download.

Log inor Register

to rate and post a comment

(3)

Reinaldo Costa

re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

May 15, 2018 12:02

Please, I NEED service manual Polydoros LX 50. how much? My e-mail: reinaldo@reitecequipamentos.com.br

Log inor Register

to rate and post a comment

(3)

Reinaldo Costa

re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

May 15, 2018 12:02

Please, I NEED service manual Polydoros LX 50. how much? My e-mail: reinaldo@reitecequipamentos.com.br

Log inor Register

to rate and post a comment

(2596)
(549)
(15)

Lenny Place

re: re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

May 16, 2018 09:20

I am sending to you

Log inor Register

to rate and post a comment

(36)
(8)

Peter Russell

re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

January 10, 2019 09:35

I should have said p.russell@medimaging.co.uk

Log inor Register

to rate and post a comment

(5)

houssein ibraheem

re: re: I NEED SERVICE MANUAL Polydoros LX 50

April 23, 2018 10:56

hello
i have error 865/016
could you please advice me,
please send me the service manual please
bme.houssein@gmail.com

Log inor Register

to rate and post a comment

(2596)
(549)
(15)

Lenny Place

re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

April 23, 2018 04:09

what system are you working on this is not a generator error code? normally a APID 16 deals with a videomed or imaging component and 865 is not showing on my list.

Log inor Register

to rate and post a comment

(5)

houssein ibraheem

re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

April 30, 2018 09:42

the table model is sireskop cx
now i facing a new error 007/035 sireskop cx
could you please send me the service manual?
my email is bme.houssein@gmail.com

Thanx
BR

Log inor Register

to rate and post a comment

(5)

houssein ibraheem

re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

May 28, 2018 10:36

good day,
these error appears on the screen
242/032
246/032
248/032

any advice please

the system is sireskope cx with polydoros lx50

Log inor Register

to rate and post a comment

(36)
(8)

Peter Russell

re: re: re: re: I NEED SERVICE MANUAL Polydoros LX 50

January 10, 2019 09:35

Hi Lenny could you send me the manual please. I have error 777 on a Tosrad with a lite generator

Log inor Register

to rate and post a comment

(462)
(63)
(1)

Mohammed Abdullah

re: I NEED SERVICE MANUAL Polydoros LX 50

January 29, 2014 06:14

i have all manuals for this system

PolyLX_Adjustment
PolyLX_Troubleshooting
PolyLX_Wiring and schematic diagram
Replacement of Parts

contact me

magmesme@gmail.com

regards

Log inor Register

to rate and post a comment

(3)

Reinaldo Costa

re: I NEED SERVICE MANUAL Polydoros LX 50

May 15, 2018 12:02

Please, I NEED service manual Polydoros LX 50. how much? My e-mail: reinaldo@reitecequipamentos.com.br

Log inor Register

to rate and post a comment

RaySafe helps you avoid unnecessary radiation

RaySafe solutions are designed to minimize the need for user interaction, bringing unprecedented simplicity & usability to the X-ray room. We're committed to establishing a radiation safety culture wherever technicians & medical staff encounter radiation.

 

advertisement

 

advertisement

 

advertisement
advertisement
O acesso e o uso deste local são sujeitos aos termos e às condições do nosso OBSERVAÇÃO LEGAL & OBSERVAÇÃO DA PRIVACIDADE
Propriedade de e proprietário DOTmeda .com, Inc. Copyright ©2001-2019 DOTmed.com, Inc.
TODOS OS DIREITOS RESERVADOS