> GE OEC 9600 C-Arm not completing Booting with Control Panel

GE OEC 9600 C-Arm not completing Booting with Control Panel

Tags: C-Arm, OEC

(164)
(29)

Amos Dantene

GE OEC 9600 C-Arm not completing Booting with Control Panel

January 03, 2018 09:11

MyGE OEC 9600 C-Arm does not complete Booting with Control Panel not displaying the exposure parameters.

Please I need help.

My email address is: adantene@gmail.com

Thanks

Log inor Register

to rate and post a comment

(93)
(10)
(1)

bashir ahmad

re: GE OEC 9600 C-Arm not completing Booting with Control Panel

January 12, 2018 03:19

your system SRAM PC card is with empty battery, it lost it setting, now you need to reload backup floppy data into sram card after replacing its battery, for sram reading and writing u need special reader.