I am faced with a problem on a Lightspeed 8.

It had a software issue with necessitated an LFC but I found that the savestate MOD was corrupted and hence could not restore system state.

I reconfigured the system manually and now trying to do a full calibration. I have successfully completed DAS Gain, Collimator and Z-slope cals but when I start the Detailed cal, the system aborts scan at Non-Bowtie Air Cals, immediately after the first exposure.
The log shows the following errors 230015013, 230015016 and 245200....

I am trying to recover the cal_db but am getting stuck here.

Log inor Register

to rate and post a comment

(202)
(25)

Karen Sarkisov

re: GE Lightspeed 8 calibration failure

September 13, 2017 09:07

Hi! you can change MOD driver, find CD driver with best quality and instal this. then reinstal SW
About errors, can you send img of this erors?