hello how are you, I'm having a mistake with a mammographer siemens mammomat C3,
when shooting the team throws error 601 at me
Anyone know what can it be?

regards
Hernan

Log inor Register

to rate and post a comment

(59)
(9)

fortronik

re: Error siemens mammomat C3

November 10, 2019 11:37



Meaning: Er 601
Missing phase or wrong phase sequence.
Description: The MAMMOMAT 2 requires the power cables to be connected in correct phase sequence to one another for generating the actuation signals for the intermediate circuit thyristors. With the power contactor NS closed (i.e. switch UZW D717.S3 must be in position ON), the correct phase position is indicated by the LED "3~" / phase sequence OK (D718.V1) Caution: If switch UZW (D717.S3) or NS (D711.S3) is in position OFF, then the line contactor NS is also open and the above LED off!
Troubleshooting: On initial startup, check that the phases are correctly connected. If this is not the case, reverse two phases.
Otherwise, first check the fuse links U1, U2 and U3 and the power line input. If these are all OK, trace the signal path from the line input through the following components: - Coils L40, L41, L42, L43
- Power line contactor
- D713 Phase detector
- D715 Electrolytic capacitor circuit
- D718 Intermediate circuit regulator
- D717 kV-regulator