BV Endura Error M351 and M208

February 11, 2016 11:47

BV Endura Error M351 and M208, Can you help me?

1 person liked this
Log inor Register

to rate and post a comment

(2431)
(486)
(13)

Lenny Place

re: BV Endura Error M351 and M208

February 11, 2016 05:53

M 351 FA_INVERTER_A LARM X-ray generator not available
Switch system off/on Call service if problem is persistent FA generator board SHA4 Inverter status signal is active. Generator is reset within a few seconds. Check the kV/mA values of the tank.


M 208 XACT_GONE Subsystem error,
system still available It's possible to continue Unexpected X-ray switch off. Check board SHA4 and the generator unit SE. Check cabling.

Log inor Register

to rate and post a comment

Dean Smith

re: re: BV Endura Error M351 and M208

February 12, 2016 09:01

U have a bad HV tank!! Replace it... Test: fails in High KV!!! Doesnt fail in low KV!!! Replace and calibrate system! U will beed BV scope and philisp Ist!!!

Regards,

Dean

Log inor Register

to rate and post a comment

odilon cardoso

re: re: re: BV Endura Error M351 and M208

February 12, 2016 01:51

When put in or more 65kV, ocurre this error, but in 40 45 50 and 60kv no problem.

Log inor Register

to rate and post a comment

(2431)
(486)
(13)

Lenny Place

re: re: re: re: BV Endura Error M351 and M208

February 12, 2016 03:40

bad mono block

Log inor Register

to rate and post a comment

(2)

Kirill Severtsov

re: re: re: re: re: BV Endura Error M351 and M208

June 07, 2018 09:42

How to fix errors M351 and M208

Log inor Register

to rate and post a comment

(2)

Kirill Severtsov

re: re: re: re: re: BV Endura Error M351 and M208

June 07, 2018 09:42

How did you fix the errors M208 and M351 of BV Endura 2.3

Log inor Register

to rate and post a comment

(407)
(33)

DMEDS

re: BV Endura Error M351 and M208

February 12, 2016 09:01

351
-
x ray generator not available Switch system off/on.
FA generator board SHA4 Inverter status signal is active. Generator is reset within a few seconds. Check the kV/mA values of the tank.

208

Subsystem error, system still available. It's possible to continue
Unexpected X ray switch off. Check board SHA4 and the
generator unit SE. Check cabling.

if you need complete service manuals contact me,

magmesme@gmail.com

Regards

Log inor Register

to rate and post a comment

(20)

lin

re: re: BV Endura Error M351 and M208

June 12, 2018 08:38

Mr i need the service manuals ,pls help me,send the manuals to the srey123@163.com,thanks

Log inor Register

to rate and post a comment

(2431)
(486)
(13)

Lenny Place

re: re: re: BV Endura Error M351 and M208

June 12, 2018 11:09

I am sending using myairbridge

Log inor Register

to rate and post a comment

(20)

lin

re: re: re: re: BV Endura Error M351 and M208

June 13, 2018 07:44

Can you send the manual by srey123@163.com ,thanks

Log inor Register

to rate and post a comment

(2431)
(486)
(13)

Lenny Place

re: re: re: re: re: BV Endura Error M351 and M208

June 14, 2018 01:44

Sending again

Log inor Register

to rate and post a comment

(20)

lin

re: re: re: re: re: re: BV Endura Error M351 and M208

June 15, 2018 07:59

Thank you !I have received the manual,the manual in the junk mail.

Log inor Register

to rate and post a comment

(2431)
(486)
(13)

Lenny Place

re: re: re: re: re: re: re: BV Endura Error M351 and M208

June 15, 2018 05:13

Hope it helps you are welcome

Log inor Register

to rate and post a comment

Your trusted and knowledgeable source for Medical Imaging Parts & Equipment

Dedicated Solutions-We provide dependable, cost effective solutions that exceed our customer's expectations & offer high quality services for your specific needs-Repairs, CT and MRI parts, staging, 24/7 Tech support, & same day shipment. Call 844-293-2057 DOTmed 100

 

 

 

O acesso e o uso deste local são sujeitos aos termos e às condições do nosso OBSERVAÇÃO LEGAL & OBSERVAÇÃO DA PRIVACIDADE
Propriedade de e proprietário DOTmeda .com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
TODOS OS DIREITOS RESERVADOS