AL03 S tar ter safety
v Alarm description:
This message is displayed when safety is active on the starter of the
anode of the X-ray tube
v Alarm cause:
Anode disconnection of X-ray tube
v Verifications:
· check A11 starter card, 22Vdc tension between X20:1 (+) and
X20:3 (-) selecting the first step of X-ray button (prep)
· check 230V tension on connector X21:01 and X23:06 of A11
starter card
· check on A5 feeder card, the passage from 0V level to 9V between
pin 15A and earth, selecting the first step of X-ray button (prep)
· check the starter OK to CPU board A7, with the change from 0V
to 13Vdc between pins X8.01 e 0V on the Mother Board A3,
pressing the first step of X-ray Button (prep.)
· check the starter access to A7 CPU card with level passage from
0V to 13Vdc between pin 15C of CPU card and 0V by selecting
the first step of X-ray button (prep)
· check if the green led on A10 starter module turns on after a
second on the first step of X-ray button (prep).
(1)
OSWELL DUBE
Villa Visitor AR30 manual
March 27, 2011 12:02
HI everyone,
I m looking for a user or service manual for a Villa visitor AR30 mobile x-ray unit.Your assitance will be greatly appreciated.
to rate and post a comment
Ngoc Vu Dinh
Re: Villa Visitor AR30 manual
March 28, 2011 10:00
Hi
I need it too. My machine shows error code: AL03.
Please send it to me if everone has.
Thanks!!!
Ngoc Vu Dinh
to rate and post a comment
Mohammed Abdullah
Re: Villa Visitor AR30 manual
March 29, 2011 07:54
Dear All,
I have the manual for Model:Vistor
not vistor AR30
If you intersted contact me on
magmesme@gmail.com
Regards
to rate and post a comment
(6)
(2)
Charbel Tadros
re: Re: Villa Visitor AR30 manual
June 06, 2014 03:36
hello,
can you send me a copy of the villa visitor service manual . an alarm 15 and 16 occurs when starting the machine .
please help
to rate and post a comment
(8)
Arturo Papa
Re: Villa Visitor AR30 manual
May 19, 2011 11:09
AL03 S tar ter safety
v Alarm description:
This message is displayed when safety is active on the starter of the
anode of the X-ray tube
v Alarm cause:
Anode disconnection of X-ray tube
v Verifications:
· check A11 starter card, 22Vdc tension between X20:1 (+) and
X20:3 (-) selecting the first step of X-ray button (prep)
· check 230V tension on connector X21:01 and X23:06 of A11
starter card
· check on A5 feeder card, the passage from 0V level to 9V between
pin 15A and earth, selecting the first step of X-ray button (prep)
· check the starter OK to CPU board A7, with the change from 0V
to 13Vdc between pins X8.01 e 0V on the Mother Board A3,
pressing the first step of X-ray Button (prep.)
· check the starter access to A7 CPU card with level passage from
0V to 13Vdc between pin 15C of CPU card and 0V by selecting
the first step of X-ray button (prep)
· check if the green led on A10 starter module turns on after a
second on the first step of X-ray button (prep).
to rate and post a comment
(6)
(2)
Charbel Tadros
re: Re: Villa Visitor AR30 manual
June 06, 2014 03:36
hello,
can you send me a copy of the villa visitor service manual . an alarm 15 and 16 occurs when starting the machine .
please help
to rate and post a comment
(19)
(1)
Jonathan Gonzales
re: re: Re: Villa Visitor AR30 manual
October 16, 2015 09:39
Good day mate, i am having an error 16 on visitor AR30? could someone know about this error?and can you send service manual?
to rate and post a comment
(59)
(3)
Jose Vallejos
Re: Villa Visitor AR30 manual
March 29, 2011 07:43
Hi for all,
I have this service and operator manual.
javv2010@live.cl
to rate and post a comment
(6)
(2)
Charbel Tadros
re: Re: Villa Visitor AR30 manual
June 06, 2014 03:36
hello,
can you send me a copy of the villa visitor service manual . an alarm 15 and 16 occurs when starting the machine .
please help
to rate and post a comment
(59)
(3)
Jose Vallejos
Re: Villa Visitor AR30 manual
March 29, 2011 07:45
Hi,
I have a service manual and operator manual for Villa Visitor AR30 Mobile.
Best regards
javv2010@live.cl
to rate and post a comment
(1)
sinisa cvetanovic
re: Re: Villa Visitor AR30 manual
October 22, 2012 05:06
hello,
i need that service manual , could you send it on my e mail address ?
to rate and post a comment
ken-ken
re: Re: Villa Visitor AR30 manual
May 21, 2015 09:01
can i have some copy of you service manual and operator manual of visitor AR30 thank you
to rate and post a comment
DMEDS
re: re: Re: Villa Visitor AR30 manual
May 25, 2015 11:39
i have it contact me,
magmesme@gmail.com
regards
to rate and post a comment
Adefemi Cole
re: Villa Visitor AR30 manual
September 20, 2019 09:22
Please can you help to Email me copy of Service Manual for Mobildrive AR30 X-ray. Thanks
to rate and post a comment