re: Siemens Siremobil Compact L service manual and error code needed
November 07, 2012 09:03
Error E7150 Slot diaphragm rotation index not found
Possible trouble: The slot diaphragm rotation index position was not found
during initialization.
Possible help: Check the index switch
Check the index switch wires.
Check the motor.
Check the motor wires.
Check the mechanical movement.
Board D1 defective.
-----------------------------
Error E7406 TV-Iris actual value out of range
Possible trouble: The actual value of the TV-Iris
is above the maximum value. (AD converter value)
Possible help: TV-Iris moved by hand ?
Wrong doserate parameters.
Actual value potentiometer defective or wrong connected.
Board D1 defective.
-----------------------------------------
Error E7410 TV-Iris limits exceeded
Possible trouble: The limits for the TV-Iris movement were exceeded.
Possible help: Check programmed parameters
Check doserate adjustment.
Check actual value potentiometer.
Check actual value potentiometer wires.
--------------------------------------------
--------------------------
Wayne M. Horsman, VP
Columbia Imaging, Inc.
re: re: re: Siemens Siremobil Compact L service manual and error code needed
December 25, 2017 08:09
Hi William and Wayne,
Can you please, please send me the error codes for siremobil compact l. I'm having trouble with error 42012 and 42024. I have been suggested to replace battery, but I cant find it.
re: re: re: re: re: Siemens Siremobil Compact L service manual and error code needed
December 26, 2017 08:34
It seems that there is a problem with ki d of battery that looks like condenser ant should be a orange colour and should be placed somewhere behind D1 board, actually on D2 board, I could not find it and the problem is that as per one technician in my region cannot save the configuration. Weird but I cannot fix the problem.
Any suggestion that can lead to fix is welcomed.
Cheers
re: re: re: Siemens Siremobil Compact L service manual and error code needed
March 09, 2016 01:48
Error 5009 is not listed, but it is probably a Sirephos error (x-ray tube)
Error E8007 Videomed DC no data transfer possible
Possible trouble: Host registered the failure of the Bright Actual telegrams of the Videomed DC.
Possible help: Host selects KV Stop automatically
Emergency operation of the system may possible.
Check power supply of the Videomed DC.
Check cables to the Videomed DC.
Videomed DC defective.
What is your email address? I will send you the error list
re: re: re: re: Siemens Siremobil Compact L service manual and error code needed
November 30, 2017 09:28
Dear Wayne, greetings. I am working on Siremobil Compact L with error 8000, any help please? Please advise, do you have troubleshooting guide with error codes? please kindly share with me.
re: re: re: re: re: Siemens Siremobil Compact L service manual and error code needed
December 01, 2017 09:18
sent more info to: ssalijoeug@gmail.com - Wayne H.
Error E8000 = Videomed DC initialization timeout
Possible trouble: Host registered timeout during initialization of the
Videomed DC.
Possible help: Check power supply of the Videomed DC.
Check cables to the Videomed DC.
Possibly Videomed DC or D1 defective.
re: re: re: re: re: re: re: Siemens Siremobil Compact L service manual and error code needed
August 23, 2019 10:09
Error E8000 Videomed DC initialization timeout
(this could be anything related to the TV Camera)
Possible trouble: Host registered timeout during initialization of the
Videomed DC.
Possible help: Check power supply of the Videomed DC.
Check cables to the Videomed DC.
Possibly Videomed DC or D1 defective.
------------
Error E4730 Memoskop CX-200 missing license
(usually means the Memoskop was replaced, but the license was not reinstalled, or maybe the USB license dongle is plugged into the wrong USB port?)
Possible trouble: During boot up of the MEMOSKOP CX-200 the license is checked
for availability and plausibility. If the license file is missing or
wrong, this error message is displayed.
Possible help: - Check the presence of the license file.
- Reinstall the license file.
Note: The license file is available on the license CD-R and also
stored on the backup CD-R (Log book).
See SIREMOBIL Compact L Service Instructions
SPR2-130.840.01.02.02 (or newer rev. level of the document)
---------
Error E7150 Slot diaphragm rotation index not found
(usually stripped teeth on the lower Slot diaphragm ring gear)
Possible trouble: The slot diaphragm rotation index position was not found
during initialization.
Possible help: Check the index switch
Check the index switch wires.
Check the motor.
Check the motor wires.
Check the mechanical movement.
Board D1 defective.
re: re: Siemens Siremobil Compact L service manual and error code needed
May 30, 2023 01:52
I replaced the D1 board, but the main system appears disconected, so I can do anything, when change the D1 board, Does it necesary do another action?, I mean, it just plug and play or it has to be load any firmware?.
Regards
Mostafa Kamel
Siemens Siremobil Compact L service manual and error code needed
November 07, 2012 10:40
Hi all.
I am having some error code on Siemens Siremobil compact L. error codes are 7150, 7406, 7410.
Does anyone have the service manual and error code list for it.
my email is mmm_kamel@yahoo.com
thanks in advance
to rate and post a comment
Ralph Frizzle
re: Siemens Siremobil Compact L service manual and error code needed
November 07, 2012 06:27
call german electronics technical support
to rate and post a comment
(1)
Ramon onelis Acosta m.
re: re: Siemens Siremobil Compact L service manual and error code needed
October 05, 2013 08:14
l need service manual siremobil compact
to rate and post a comment
Horia Ioan
re: re: re: Siemens Siremobil Compact L service manual and error code needed
May 31, 2019 08:40
Could somebody help me with service manual siremobil compact l. Thank you
to rate and post a comment
(5)
Matias Roldan
re: re: re: re: Siemens Siremobil Compact L service manual and error code needed
June 03, 2019 09:22
write me danielrkoorton@gmail.com
to rate and post a comment
(1)
Sheyin Bikai
re: re: Siemens Siremobil Compact L service manual and error code needed
July 15, 2014 04:19
Am having an error 9702 can you please suggest what to do please.
to rate and post a comment
(47)
(28)
Wayne Horsman
re: re: re: Siemens Siremobil Compact L service manual and error code needed
July 16, 2014 09:09
Error E9702 = Diamentor Wrong test value
Possible trouble: Host reports wrong test value for Diamentor.
Possible help: Check +15V power supply.
Adjust Diamentor.
Check parameters for Diamentor for correctness.
Diamentor defective.
to rate and post a comment
bilal master
re: re: re: re: Siemens Siremobil Compact L service manual and error code needed
October 12, 2018 08:18
Hi all friends
Please help me
siemens siremobil 4 h service manual please send me
bilalmaster9@gmail.com
to rate and post a comment
(47)
(28)
Wayne Horsman
re: re: re: Siemens Siremobil Compact L service manual and error code needed
July 16, 2014 09:10
Error E9702 => Diamentor Wrong test value
Possible trouble: Host reports wrong test value for Diamentor.
Possible help: Check +15V power supply.
Adjust Diamentor.
Check parameters for Diamentor for correctness.
Diamentor defective.
to rate and post a comment
bilal master
re: re: Siemens Siremobil Compact L service manual and error code needed
October 12, 2018 08:16
Hi all friends
Please help me
siemens siremobil 4 h service manual please send me
bilalmaster9@gmail.com
to rate and post a comment
(47)
(28)
Wayne Horsman
re: Siemens Siremobil Compact L service manual and error code needed
November 07, 2012 09:03
Error E7150 Slot diaphragm rotation index not found
Possible trouble: The slot diaphragm rotation index position was not found
during initialization.
Possible help: Check the index switch
Check the index switch wires.
Check the motor.
Check the motor wires.
Check the mechanical movement.
Board D1 defective.
-----------------------------
Error E7406 TV-Iris actual value out of range
Possible trouble: The actual value of the TV-Iris
is above the maximum value. (AD converter value)
Possible help: TV-Iris moved by hand ?
Wrong doserate parameters.
Actual value potentiometer defective or wrong connected.
Board D1 defective.
-----------------------------------------
Error E7410 TV-Iris limits exceeded
Possible trouble: The limits for the TV-Iris movement were exceeded.
Possible help: Check programmed parameters
Check doserate adjustment.
Check actual value potentiometer.
Check actual value potentiometer wires.
--------------------------------------------
--------------------------
Wayne M. Horsman, VP
Columbia Imaging, Inc.
wayne@columbiaimaging.com
--------------------------
to rate and post a comment
(2)
william oliver
re: re: Siemens Siremobil Compact L service manual and error code needed
October 26, 2014 08:20
Hi Wayne Horsman .
I'm having an error code on error codes Siemens SIREMOBIL COMPACT L. are 8000, 2006, 4021, 4024.
Does anyone have the service manual, list of error codes for it or some solution.
my email is williamoliveira7@gmail.com
Thanks in advance
to rate and post a comment
(1)
Albion Bikliqi
re: re: re: Siemens Siremobil Compact L service manual and error code needed
December 25, 2017 08:09
Hi William and Wayne,
Can you please, please send me the error codes for siremobil compact l. I'm having trouble with error 42012 and 42024. I have been suggested to replace battery, but I cant find it.
to rate and post a comment
(47)
(28)
Wayne Horsman
re: re: re: re: Siemens Siremobil Compact L service manual and error code needed
December 26, 2017 08:42
what battery? also, I don't recognize error 42012 and 42024
to rate and post a comment
(1)
Albion Bikliqi
re: re: re: re: re: Siemens Siremobil Compact L service manual and error code needed
December 26, 2017 08:34
It seems that there is a problem with ki d of battery that looks like condenser ant should be a orange colour and should be placed somewhere behind D1 board, actually on D2 board, I could not find it and the problem is that as per one technician in my region cannot save the configuration. Weird but I cannot fix the problem.
Any suggestion that can lead to fix is welcomed.
Cheers
to rate and post a comment
(1)
Albion Bikliqi
re: re: re: Siemens Siremobil Compact L service manual and error code needed
December 25, 2017 08:09
My email is albionbikliqi@gmail.com
Thanks in advance and Marry Christmas to all
to rate and post a comment
(1)
Albion Bikliqi
re: re: re: Siemens Siremobil Compact L service manual and error code needed
December 25, 2017 08:09
My email is albionbikliqi@gmail.com
Thanks in advance and Marry Christmas to all
to rate and post a comment
(2)
(8)
Jairo Manzaneda
re: re: Siemens Siremobil Compact L service manual and error code needed
March 09, 2016 09:20
i have a systems like this with error code 8007 and 5009 can you help me
to rate and post a comment
(47)
(28)
Wayne Horsman
re: re: re: Siemens Siremobil Compact L service manual and error code needed
March 09, 2016 01:48
Error 5009 is not listed, but it is probably a Sirephos error (x-ray tube)
Error E8007 Videomed DC no data transfer possible
Possible trouble: Host registered the failure of the Bright Actual telegrams of the Videomed DC.
Possible help: Host selects KV Stop automatically
Emergency operation of the system may possible.
Check power supply of the Videomed DC.
Check cables to the Videomed DC.
Videomed DC defective.
What is your email address? I will send you the error list
to rate and post a comment
(32)
(9)
joe ssali
re: re: re: re: Siemens Siremobil Compact L service manual and error code needed
November 30, 2017 09:28
Dear Wayne, greetings. I am working on Siremobil Compact L with error 8000, any help please? Please advise, do you have troubleshooting guide with error codes? please kindly share with me.
to rate and post a comment
(47)
(28)
Wayne Horsman
re: re: re: re: re: Siemens Siremobil Compact L service manual and error code needed
December 01, 2017 09:18
sent more info to: ssalijoeug@gmail.com - Wayne H.
Error E8000 = Videomed DC initialization timeout
Possible trouble: Host registered timeout during initialization of the
Videomed DC.
Possible help: Check power supply of the Videomed DC.
Check cables to the Videomed DC.
Possibly Videomed DC or D1 defective.
to rate and post a comment
ALEJANDRO MORENO
re: re: re: re: re: re: Siemens Siremobil Compact L service manual and error code needed
August 22, 2019 09:24
Hello Wayne, I have problems with error 8000, 4730 and 7150 can you tell me about this errors? Have you a service manual?
to rate and post a comment
(47)
(28)
Wayne Horsman
re: re: re: re: re: re: re: Siemens Siremobil Compact L service manual and error code needed
August 23, 2019 10:09
Error E8000 Videomed DC initialization timeout
(this could be anything related to the TV Camera)
Possible trouble: Host registered timeout during initialization of the
Videomed DC.
Possible help: Check power supply of the Videomed DC.
Check cables to the Videomed DC.
Possibly Videomed DC or D1 defective.
------------
Error E4730 Memoskop CX-200 missing license
(usually means the Memoskop was replaced, but the license was not reinstalled, or maybe the USB license dongle is plugged into the wrong USB port?)
Possible trouble: During boot up of the MEMOSKOP CX-200 the license is checked
for availability and plausibility. If the license file is missing or
wrong, this error message is displayed.
Possible help: - Check the presence of the license file.
- Reinstall the license file.
Note: The license file is available on the license CD-R and also
stored on the backup CD-R (Log book).
See SIREMOBIL Compact L Service Instructions
SPR2-130.840.01.02.02 (or newer rev. level of the document)
---------
Error E7150 Slot diaphragm rotation index not found
(usually stripped teeth on the lower Slot diaphragm ring gear)
Possible trouble: The slot diaphragm rotation index position was not found
during initialization.
Possible help: Check the index switch
Check the index switch wires.
Check the motor.
Check the motor wires.
Check the mechanical movement.
Board D1 defective.
to rate and post a comment
(3)
Walter Gillekens
re: re: Siemens Siremobil Compact L service manual and error code needed
July 04, 2017 12:43
Hi Wayne,
Can you please send me the overview of all Siremobil Compact L error codes (gwmedical81@gmail.com) Thanks for your help.
Best regards,
Walter
to rate and post a comment
Valentina Rosano
re: re: Siemens Siremobil Compact L service manual and error code needed
June 08, 2018 09:38
Hello,
do you have any idea which fault could be the 7400 error code?Can you help me?
Thanks in advance,
Valentina Rosano
X Medical srl
to rate and post a comment
(3)
(1)
Lobsang Ortiz
re: re: Siemens Siremobil Compact L service manual and error code needed
May 30, 2023 01:52
I replaced the D1 board, but the main system appears disconected, so I can do anything, when change the D1 board, Does it necesary do another action?, I mean, it just plug and play or it has to be load any firmware?.
Regards
to rate and post a comment
Mohammed Abdullah
re: Siemens Siremobil Compact L service manual and error code needed
November 07, 2012 09:03
Kindly contact me
magmesme@gmail.com
to rate and post a comment
(1)
Ron Levchuk
re: re: Siemens Siremobil Compact L service manual and error code needed
July 24, 2013 06:26
Hi, Can you email me the service manual.
rlevchuk@gmail.com
Thanks, Ron
to rate and post a comment
(1)
Sheyin Bikai
re: Siemens Siremobil Compact L service manual and error code needed
July 15, 2014 04:19
Hello Mostafa, have you gotten the service manual for siremobil compact L?Am also having an error 9702. Can you help in anyway please.
to rate and post a comment
Georgy Smith
re: re: Siemens Siremobil Compact L service manual and error code needed
July 16, 2014 09:09
write to me tomography78@gmail.com
to rate and post a comment
zurabi giorgadze
re: Siemens Siremobil Compact L service manual and error code needed
November 06, 2019 10:00
Could help me with service manual siremobil compact .
error codes 2001.2006,0001
Thank you
to rate and post a comment