Hello!
7408
Explanation: a) TV-Iris failure.
TV-Iris: Unexpected movement
The TV-Iris movement was not activated, but the system detected a movement.
Action: a1) Check TV-Iris parameters.
a1) Check actual value potentiometer.
a2) Check wires.
a3) Check TV-Iris (I.I. optical system).
a4) The Main System D1 might be defect.
7409
Explanation: a) TV-Iris failure.
TV-Iris: Wrong position
If the failure 7400 to 7408 were detected,
this failure is displayed after each radiation release.
User warning:
The user must delete this failure
before radiation can be released again.
Action: a1) Check for the event log id with the source ASP_MSP 74*
(1)
garanich oleg
siemens arcadis tv-iris failure
July 20, 2021 10:50
Hello
siemens arcadis gives tv-iris failure ASP_E_7408 or
7409
sometimes after reboot everything is fine
Please tell me what to watch
to rate and post a comment
(16)
(17)
Doscar Ernesto Perez Pereyra
re: siemens arcadis tv-iris failure
July 21, 2021 09:30
You need to enter to Error log for watch these error but normally fixed with the camera replace
to rate and post a comment
(72)
(5)
Marko Shevchuk
re: siemens arcadis tv-iris failure
July 21, 2021 09:31
Hello!
7408
Explanation: a) TV-Iris failure.
TV-Iris: Unexpected movement
The TV-Iris movement was not activated, but the system detected a movement.
Action: a1) Check TV-Iris parameters.
a1) Check actual value potentiometer.
a2) Check wires.
a3) Check TV-Iris (I.I. optical system).
a4) The Main System D1 might be defect.
7409
Explanation: a) TV-Iris failure.
TV-Iris: Wrong position
If the failure 7400 to 7408 were detected,
this failure is displayed after each radiation release.
User warning:
The user must delete this failure
before radiation can be released again.
Action: a1) Check for the event log id with the source ASP_MSP 74*
to rate and post a comment