PROBLEM IN G.E SENOGRAPHE 800T. E12 CHEKSUM ERROR 071/081
December 31, 2011 11:55
Hello, I have problems with a GE senographe 800T, showing the following error: E12 Checksum Error 071/081. Someone I could help. My email is: jose.ig.herrera@gmail.com.
Re: PROBLEM IN G.E SENOGRAPHE 800T. E12 CHEKSUM ERROR 071/081
December 31, 2011 05:34
Error 071/081 Generator: checksum error on battery saved memory
GENERATOR CPU/: This error has at least three causes:
1 a) After a calibration in the install menu a checksum has not been performed to validate the new data.
1 b) One parameter (or more) has been changed manually by mistake by the FE. Do not perform the checksum, and try to find what parameter has been changed.
2) The lithium battery is dead, or the jumper X2 on the CPU generator is not at the correct location. This can happen at the time of the replacement of the CPU board. The reason for that is that when the DMR is powered, the lithium battery is not needed. see SCH. 800PL3 400PL3 page 9/10.
3) When performing a reload the CPU with save parameters.
(16)
(1)
JOSE HERRERA
PROBLEM IN G.E SENOGRAPHE 800T. E12 CHEKSUM ERROR 071/081
December 31, 2011 11:55
Hello, I have problems with a GE senographe 800T, showing the following error: E12 Checksum Error 071/081. Someone I could help. My email is: jose.ig.herrera@gmail.com.
Thanks and happy new year.
José I
to rate and post a comment
(6)
Jaime Diaz
Re: PROBLEM IN G.E SENOGRAPHE 800T. E12 CHEKSUM ERROR 071/081
December 31, 2011 05:34
Error 071/081 Generator: checksum error on battery saved memory
GENERATOR CPU/: This error has at least three causes:
1 a) After a calibration in the install menu a checksum has not been performed to validate the new data.
1 b) One parameter (or more) has been changed manually by mistake by the FE. Do not perform the checksum, and try to find what parameter has been changed.
2) The lithium battery is dead, or the jumper X2 on the CPU generator is not at the correct location. This can happen at the time of the replacement of the CPU board. The reason for that is that when the DMR is powered, the lithium battery is not needed. see SCH. 800PL3 400PL3 page 9/10.
3) When performing a reload the CPU with save parameters.
to rate and post a comment
(5)
Luis DeLeon Grimaldi
re: Re: PROBLEM IN G.E SENOGRAPHE 800T. E12 CHEKSUM ERROR 071/081
July 12, 2012 01:46
Hello Jaime, How can I reload the CPU? Thank you for your Help, Luis
to rate and post a comment
ruby mole kr
re: re: Re: PROBLEM IN G.E SENOGRAPHE 800T. E12 CHEKSUM ERROR 071/081
March 17, 2014 01:40
i have the same problem in my machine .if you pls help how an i reload the cpu
to rate and post a comment
ruby mole kr
re: re: Re: PROBLEM IN G.E SENOGRAPHE 800T. E12 CHEKSUM ERROR 071/081
March 17, 2014 01:41
i also have the same problem how can i reload the cpu would some one help me pls
to rate and post a comment
(2)
karen alberto garcia salazar
re: re: re: Re: PROBLEM IN G.E SENOGRAPHE 800T. E12 CHEKSUM ERROR 071/081
January 22, 2018 09:59
I also have the problem with the card I will have to change the card. I think the eprom of the card deleted the information
to rate and post a comment
(2)
karen alberto garcia salazar
re: re: Re: PROBLEM IN G.E SENOGRAPHE 800T. E12 CHEKSUM ERROR 071/081
January 22, 2018 09:59
Jaime I did all that you say and it remains the same you have to replace the entire card. I can only solve the team
to rate and post a comment
Jean baptiste Ndah
re: re: re: Re: PROBLEM IN G.E SENOGRAPHE 800T. E12 CHEKSUM ERROR 071/081
April 14, 2019 04:08
Hello
I am jean i am having this message on the console of my GE senographe DMR Plus: E12 Checksum error (gen) 071/081
Can someone help me
to rate and post a comment
(2)
alen kadic
re: re: re: re: Re: PROBLEM IN G.E SENOGRAPHE 800T. E12 CHEKSUM ERROR 071/081
November 02, 2022 11:46
Hello did you solve this problem I have the same
to rate and post a comment