re: Digital Mammogram Senographe Essential GE error code
November 18, 2020 09:54
162/007 INHIBIT GRID NODE NOT
RESPONDING
Possible causes:
• The CAN cables W503 and/or W504 on the breast
support locking mechanism are either faulty or
badly connected.
• The grid node did not send the heatbeat signal.
Between two motions, the grid node sends a Heartbeat
signal every second to Gantry CPU. If not, the
Gantry CPU creates this error. During lift motion,
the grid node sends an actual value message regularly
in addition to the heartbeat. If not, the Gantry
CPU creates this error.
Proceed as follows:
1. Reboot the system from the X-ray Console.
2. If the problem persists, check the CAN cables
W503 and W504 on the Breast Support Locking
Mechanism Board. If the error persists once you
checked that the connections are fine, the CAN
cable W503 and/or W504 is faulty and must be
changed
SPECT, PET, CT and PET/CT equipment sales and service from OEM certified FSEs. Maintenance, repair, installation, relocation, and parts available with 24/7 support. 1-866-682-6331 or email sales@TTGimagingsolutions.com.
Hemn Rahman
Digital Mammogram Senographe Essential GE error code
November 17, 2020 11:53
Our system shows an error code of ''162007'', anyone knows what the issue is? Thanks in advance.
to rate and post a comment
(9)
Francesco Didonato
re: Digital Mammogram Senographe Essential GE error code
November 18, 2020 09:54
162/007 INHIBIT GRID NODE NOT
RESPONDING
Possible causes:
• The CAN cables W503 and/or W504 on the breast
support locking mechanism are either faulty or
badly connected.
• The grid node did not send the heatbeat signal.
Between two motions, the grid node sends a Heartbeat
signal every second to Gantry CPU. If not, the
Gantry CPU creates this error. During lift motion,
the grid node sends an actual value message regularly
in addition to the heartbeat. If not, the Gantry
CPU creates this error.
Proceed as follows:
1. Reboot the system from the X-ray Console.
2. If the problem persists, check the CAN cables
W503 and W504 on the Breast Support Locking
Mechanism Board. If the error persists once you
checked that the connections are fine, the CAN
cable W503 and/or W504 is faulty and must be
changed
to rate and post a comment
(9)
Francesco Didonato
re: Digital Mammogram Senographe Essential GE error code
November 18, 2020 09:54
or board grid node broken
to rate and post a comment