CPI CMP 200 50 KW X RAY GENERATOR

September 10, 2019 11:57

i have CMP CPI 200 X RAY GENERATOR .
when i do calibration kv change . but mA cant be change during calibration
what kind of problem it have.
does any one have UVPROM XFILE.
so that i burn it . may be possible that problem in file. does any one have the right solution .
waiting for response .

Log inor Register

to rate and post a comment

(5)
(7)

SuperJoe

re: CPI CMP 200 50 KW X RAY GENERATOR

September 29, 2019 04:37

First of all select the correct tube in the library of available tubes according to the configuration of that generator.

Then execute the generator calibration procedure for that tube (always starting with coarse focus). Just keep the trigger fully depressed and not so firm (so you don't get tired), then watch the generator go up the filament current for each KV station. Finally it will say "Calibration successfully completed". Then continue with the fine focus. If there are error messages during this process you should check according to the "troubleshooting" section of the service manual. The most common is an X-ray tube with problems (metallized and / or gasified which will generate an electric arc during the calibration process).

Good luck and be careful.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs.

Medilab Global provides high-quality pre-owned imaging equipment, parts, and service to the healthcare industry worldwide. Our number 1 position is to provide excellent customer service.. 305-234-0084 | Email: info@medilabglobal.com | Hablamos Español

DOTmed Certified DOTmed 100 stats
advertisement

Próximos Eventos de Subasta

NorthWest Supply
TIMED
Apr 08, 2024 (99 lots)