Your Email:
Walid Abu Farha
Can we install Dura tube for Siemens sensation 16 CT instead of
October 02, 2019 10:12
Can we install Dura tube for Siemens sensation 16 CT instead of Straton Tube
to rate and post a comment
Eric Hauk
re: Can we install Dura tube for Siemens sensation 16 CT instead of
October 03, 2019 10:11
You can but takes some modifications to the Cbox i believe. It’s been a long time since I tackled that project
Lin Dong
Yes, but you need to change the HV system,and softwrae. big job
Gabriel Cool
October 03, 2019 10:17
No
HAssadi
October 03, 2019 03:24
Yes you can but it requires alot of work. The gantry weights must be swapped out with a set from an akron q or dunlee tube.
re: re: Can we install Dura tube for Siemens sensation 16 CT instead of
October 04, 2019 10:11
No he can't. It's will be easy to change totally gentry. He need ti change colimatir, ebox, generator, DOM. And how he can do this?
Your comprehensive contrast injector solution for 10 years to include: on-site service, telephone technical support, depot repair with loaner options, parts sales, injector sales, injector install & de-install, Level 2 & beyond training.
(3)
Walid Abu Farha
Can we install Dura tube for Siemens sensation 16 CT instead of
October 02, 2019 10:12
Can we install Dura tube for Siemens sensation 16 CT instead of Straton Tube
to rate and post a comment
(4)
Eric Hauk
re: Can we install Dura tube for Siemens sensation 16 CT instead of
October 03, 2019 10:11
You can but takes some modifications to the Cbox i believe. It’s been a long time since I tackled that project
to rate and post a comment
(14)
(4)
Lin Dong
re: Can we install Dura tube for Siemens sensation 16 CT instead of
October 03, 2019 10:11
Yes, but you need to change the HV system,and softwrae. big job
to rate and post a comment
(67)
Gabriel Cool
re: Can we install Dura tube for Siemens sensation 16 CT instead of
October 03, 2019 10:17
No
to rate and post a comment
HAssadi
re: Can we install Dura tube for Siemens sensation 16 CT instead of
October 03, 2019 03:24
Yes you can but it requires alot of work. The gantry weights must be swapped out with a set from an akron q or dunlee tube.
to rate and post a comment
(67)
Gabriel Cool
re: re: Can we install Dura tube for Siemens sensation 16 CT instead of
October 04, 2019 10:11
No he can't. It's will be easy to change totally gentry. He need ti change colimatir, ebox, generator, DOM. And how he can do this?
to rate and post a comment